Author Topic: Fluke 287 Inaccurate on 500Ohms Range  (Read 701 times)

0 Members and 1 Guest are viewing this topic.

Offline vladsolTopic starter

  • Contributor
  • Posts: 48
  • Country: 00
Fluke 287 Inaccurate on 500Ohms Range
« on: October 17, 2024, 03:31:29 pm »
Hi all.

According to the datasheet, the Fluke 28* has a short-circuit current of 1mA when measuring resistance at the 500-ohm range. Can someone check whether this value is exactly 1 mA or if it differs?

I have a Fluke 287 with a significant deviation when measuring resistance below 500 ohms (in the 500-ohm range). Calibration of the multimeter does not fix the issue. On the 5K range and higher, this problem does not occur. I suspect the issue may be with the SL10327 chip, specifically in its current source.

What's interesting: the deviation from the actual value decreases at the edges of the range. For example (resistor value, reading, difference):
10Ω - 10.54Ω (+0.54Ω)
20Ω - 21.06Ω (+1.06Ω)
30Ω - 31.56Ω (+ 1.56Ω)
40Ω - 42.03Ω (+ 2.03Ω)
50Ω - 52.49Ω (+ 2.49Ω)
60Ω - 62.91Ω (+ 2.91Ω)
70Ω - 73.32Ω (+ 3.32Ω)
80Ω - 83.70Ω (+ 3.70Ω)
90Ω - 94.06Ω (+4.06Ω)
100Ω - 104.14Ω (+4.1Ω)
200Ω - 205.95Ω (+5.95Ω)
300Ω - 305.43Ω (+5.43Ω)
400Ω - 402.64Ω (+2.64Ω)
500Ω - 497.61Ω (-2.39Ω)

In the 5K range, the same resistors show:
100Ω - 0.1000K
200Ω - 0.2000K
300Ω - 0.3000K
400Ω - 0.4000K
500Ω - 0.5000K
« Last Edit: October 17, 2024, 05:45:37 pm by vladsol »
 

Offline gamalot

  • Super Contributor
  • ***
  • Posts: 1389
  • Country: au
  • Correct my English
    • Youtube
Re: Fluke 287 Inaccurate on 500Ohms Range
« Reply #1 on: October 17, 2024, 06:56:07 pm »
I don't think a precise value of current is necessary for resistance measurement, it just needs to be stable.
I'm a poet, I didn't even know it. |  https://youtube.com/@gamalot | https://github.com/gamalot
 
The following users thanked this post: vladsol

Offline vladsolTopic starter

  • Contributor
  • Posts: 48
  • Country: 00
Re: Fluke 287 Inaccurate on 500Ohms Range
« Reply #2 on: October 17, 2024, 09:14:11 pm »
I don't think a precise value of current is necessary for resistance measurement, it just needs to be stable.
Thank you for the photo! Unfortunately, in my case, this current is less than 1mA - around 0.94mA.
Without an accurate current source (or without current feedback in the resistance measurement circuit), it won't be possible to measure the resistance precisely, since the resistance is measured based on the voltage drop across the terminals of the multimeter (or across the X resistance) with known current.

« Last Edit: October 19, 2024, 01:34:31 pm by vladsol »
 

Offline kblue

  • Contributor
  • Posts: 39
  • Country: de
Re: Fluke 287 Inaccurate on 500Ohms Range
« Reply #3 on: October 18, 2024, 01:39:38 am »
1mA for me too. Guess this should be expected? I wonder whats wrong with your unit. Hm.

2405483-0
 
The following users thanked this post: vladsol

Offline beenai2018

  • Contributor
  • Posts: 30
  • Country: us
Re: Fluke 287 Inaccurate on 500Ohms Range
« Reply #4 on: October 18, 2024, 08:35:03 pm »
If the current was lower than 1mA wouldn't the resistance indicate be lower than normal ?

In my mind, I's seeing somewhat of a linear line to 100ohm  and then all over the place to 500.
« Last Edit: October 18, 2024, 08:41:58 pm by beenai2018 »
 

Offline vladsolTopic starter

  • Contributor
  • Posts: 48
  • Country: 00
Re: Fluke 287 Inaccurate on 500Ohms Range
« Reply #5 on: October 19, 2024, 07:36:41 am »
If the current was lower than 1mA wouldn't the resistance indicate be lower than normal ?

In my mind, I's seeing somewhat of a linear line to 100ohm  and then all over the place to 500.


Before calibration, the situation was a bit different :) After calibration, it got better, but the issue still remains. Before calibration, the resistance was indeed lower (the resistance of a 100Ω (0.01%) resistor was measured as ~94Ω).
I ordered a new SL10327 chip (checked the clamping diodes on the ISRC line, and they are OK), I will try to replace it... If the current value is 1mA, I will send it for calibration again.
« Last Edit: October 19, 2024, 08:15:18 am by vladsol »
 

Offline armandine2

  • Frequent Contributor
  • **
  • Posts: 715
  • Country: gb
Re: Fluke 287 Inaccurate on 500Ohms Range
« Reply #6 on: October 19, 2024, 01:21:15 pm »
I don't think a precise value of current is necessary for resistance measurement, it just needs to be stable.

I wonder - may be?
In a closed society where everybody's guilty, the only crime is getting caught - Hunter S Thompson
 
The following users thanked this post: vladsol

Offline floobydust

  • Super Contributor
  • ***
  • Posts: 7530
  • Country: ca
Re: Fluke 287 Inaccurate on 500Ohms Range
« Reply #7 on: October 19, 2024, 06:35:47 pm »
Have you confirmed the Ohms-source protection circuit is not damaged - the PTC and 1k surge resistor, the clamp diodes etc. are OK?
If you take the meter apart and heat up those parts, or the SL10327 - that can flush out the source of stray leakage current or high resistance.

It could be a bad CMOS switch inside the SL10327 or the Ohms compliance voltage on the 500 ohm range is flushing out a problem.
 

Offline vladsolTopic starter

  • Contributor
  • Posts: 48
  • Country: 00
Re: Fluke 287 Inaccurate on 500Ohms Range
« Reply #8 on: October 19, 2024, 07:41:29 pm »
Have you confirmed the Ohms-source protection circuit is not damaged - the PTC and 1k surge resistor, the clamp diodes etc. are OK?
Yup, I have checked the current source protection and the sense line protection.
The supply voltage of the SL10327 is ok, and the current-setting resistors R173, R71, R72 are fine as well. C14 and C17 seem alive too :)
 