Author Topic: RadioShack FET VOM repair has me stumped  (Read 552 times)

blue_lateral and 4 Guests are viewing this topic.

Online RichardMTopic starter

  • Frequent Contributor
  • **
  • Posts: 251
  • Country: au
RadioShack FET VOM repair has me stumped
« on: October 10, 2024, 11:24:15 am »
Hi All

I have an old Micronta 22-208 analog FET voltmeter that works perfectly except for resistances on the highest 1M range which measure high by about a factor of 2. I cannot for the life of me work out why.

I have checked the 9.9M resistor R19 out of circuit and it works fine. The S1 switch “fingers” also seem to make contact with the right pads on the dial pcb.

I thought I would try to measure resistance between the COM terminal and the battery positive terminal whilst rotating the S1 dial between the resistance ranges. I measure the correct values of about 10ohms, 100ohms, 10kohms, 100kohm on the lowest ranges but only 2.17Mohms on the 1M range?

Any help appreciated. Scratching my head at the moment.

Richard

 

Offline xavier60

  • Super Contributor
  • ***
  • Posts: 2925
  • Country: au
Re: RadioShack FET VOM repair has me stumped
« Reply #1 on: October 10, 2024, 12:15:56 pm »
Can R19 be measured in circuit? Maybe a conductive track has formed between the contact pads of the switch.
HP 54645A dso, Fluke 87V dmm,  Agilent U8002A psu,  FY6600 function gen,  Brymen BM857S, HAKKO FM-204, New! HAKKO FX-971.
 

Offline MathWizard

  • Super Contributor
  • ***
  • Posts: 1613
  • Country: ca
Re: RadioShack FET VOM repair has me stumped
« Reply #2 on: October 10, 2024, 07:36:21 pm »
It's a cool circuit, I started to make the circuit, but IDK if LTSpice has any dual JFET's like that. But later I want to use regular JFET's, and alter whatever, to match the input/output range of the JFET's as a diff-amp.

But yeah I've never tried to make a pure meter for V or A or Ohms. I would also like to try making a simple deflection meter, with some coil, even just wrapped around a compass.

One of these days...
 

Offline floobydust

  • Super Contributor
  • ***
  • Posts: 7440
  • Country: ca
Re: RadioShack FET VOM repair has me stumped
« Reply #3 on: October 10, 2024, 08:24:14 pm »
Measuring DCV, is there symmetry in readings between switch S3 in Polarity(+) or (-) ?
i.e. DCV inject almost a volt on say the 1V range, compare what it reads with Polarity(+) verses leads swapped and Polarity(-).
Just to see if one JFET side has some excess leakage current on the gate circuit cause a loading error.
If that is the situation, it would explain R19 appearing low value and Rx1MEG reading low as well.
 

Online RichardMTopic starter

  • Frequent Contributor
  • **
  • Posts: 251
  • Country: au
Re: RadioShack FET VOM repair has me stumped
« Reply #4 on: October 10, 2024, 09:15:23 pm »
Thanks for the replies.

In circuit, R19 measures 9.9M right around the dial EXCEPT when on the 1M range where it measures about 4.8M ?

The DC voltages are symmetric when changing the polarity switch which is a relief as the jfet may be fine?

If I measure between each of the pads for the ranges and the inner arc shaped pad on the switch (one with the 1.5V) I get around 3.68M ohms. If this is in parallel with the 9.8M resister then that would give me the 2.17M I am reading between COM and the battery positive terminal. The 3.68M in parallel with the other resistors won’t make much of a difference. Just some random thoughts from probing around

Richard
 

Online RichardMTopic starter

  • Frequent Contributor
  • **
  • Posts: 251
  • Country: au
Re: RadioShack FET VOM repair has me stumped
« Reply #5 on: October 10, 2024, 09:40:01 pm »
Switching the polarity switch from +DC to -DC the measurement across R19 changes from 2.17M to 3.3M ?
 

Offline xavier60

  • Super Contributor
  • ***
  • Posts: 2925
  • Country: au
Re: RadioShack FET VOM repair has me stumped
« Reply #6 on: October 10, 2024, 11:23:14 pm »
Has the 1.5V cell ever leaked. The electrolyte can wick along stranded wires and cause contamination at the far end.
HP 54645A dso, Fluke 87V dmm,  Agilent U8002A psu,  FY6600 function gen,  Brymen BM857S, HAKKO FM-204, New! HAKKO FX-971.
 

Online RichardMTopic starter

  • Frequent Contributor
  • **
  • Posts: 251
  • Country: au
Re: RadioShack FET VOM repair has me stumped
« Reply #7 on: October 10, 2024, 11:59:52 pm »
It is a new 1.5v cell

Also the R19 measurements do not change when the battery is removed?
 

Offline xavier60

  • Super Contributor
  • ***
  • Posts: 2925
  • Country: au
Re: RadioShack FET VOM repair has me stumped
« Reply #8 on: Yesterday at 12:24:25 am »
A BTW for anyone suspecting electrolyte contamination, it usually causes small voltages to be measured in affected circuitry even though all external power sources have been removed.
HP 54645A dso, Fluke 87V dmm,  Agilent U8002A psu,  FY6600 function gen,  Brymen BM857S, HAKKO FM-204, New! HAKKO FX-971.
 

Offline floobydust

  • Super Contributor
  • ***
  • Posts: 7440
  • Country: ca
Re: RadioShack FET VOM repair has me stumped
« Reply #9 on: Yesterday at 01:03:10 am »
If you look at the schematic, the entire metering section, JFET, 9V battery is floating and isolated from the rest of the multimeter (divider strings, shunts, rotary switch etc.)
Only Points G, A, E between them.

It seems to be around 2.7MEG worth of leakage yet R25 is 5MEG so what's with that. But changing the Polarity switch and getting 3.3MEG for a different leakage resistance of near 5MEG now seems baffling as well.
OP what type of multimeter are you using, to know its Ohms voltage and test current etc. if it is throwing things off.
Point "G" should be high resistance unless it's pushing enough to bias the JFET junction or (reverse) turn on zener D1. Check R25 is 5MEG.

Ensure there is no lithium grease used on the rotary switch, that is a no-no. Lithium is a salt and can be conductive lube. I would clean the PC board with Q-tips and IPA, get rid of old flux, soot, carbon etc.
I think it's hard to tell if a component's value is low or it's contamination.
 

Online schmitt trigger

  • Super Contributor
  • ***
  • Posts: 2334
  • Country: mx
Re: RadioShack FET VOM repair has me stumped
« Reply #10 on: Yesterday at 01:21:54 am »
I also had this same FET-VOM.
I loved it, but for a weird reason, the plastic pieces started to rot away. Disintegrating
into plastic dust.
First to go was the back case, but I could somehow mend it. That is, until the selector switch simply fell apart.
 

Online RichardMTopic starter

  • Frequent Contributor
  • **
  • Posts: 251
  • Country: au
Re: RadioShack FET VOM repair has me stumped
« Reply #11 on: Yesterday at 03:00:34 am »
I have an assortment of meters:) but have been using a Fluke 87V, HP3468a and HP3478a. I also have 2 HP-410C I could try.

I have cleaned the rotary switch contacts well with IPA and a light sand with a fibreglass pen but will look at cleaning the meter more widely. Maybe the polarity switch is the culprit? I will report any changes.

Cheers
 

Online RichardMTopic starter

  • Frequent Contributor
  • **
  • Posts: 251
  • Country: au
Re: RadioShack FET VOM repair has me stumped
« Reply #12 on: Today at 12:21:55 am »
Ok, so it is not the polarity switch. I desoldered it and hard wired the polarity and I see the same behaviour on the 1M range.
 