Hi
May I first say that it does not help and adds to the confusion when what should be two separate threads become merged into one.
I intend addressing the problems with the one in France only in this posting, given, that it appears to be the simplest/quickest one to fault find.
The printed service manual that I have is dated 8th March 1999, exactly the same as the online version.
Errors in operators and service manuals are not unheard of, this covers all manufacturers, not just Marconi. Your brain has to be kept active somehow.
Having re read the entire posting very carefully and if the information provided is accurate the initial problem with the one in France is clear. It is stated that after a few seconds the 10MHz standard output signal on the rear of the instrument goes missing. This is the first fault to be addressed, without a fully working 10MHz internal standard the instrument cannot possibly work correctly.
There are two types of 10MHz standard used in the 2023/2025 range of Marconi signal generators, a TCXO one made by HY-Q and a OCXO one made by NDK (option 4 on serial plate). I have yet to see a faulty NDK standard but this is certainly not the case with the HY-Q standard which do fail. BUT it must be made clear that the problem may not be one the 10MHz standards at all, they rely on the instruments power supply and the support circuitry on the AB1 control board all of which must also be checked and be fully functional.
Check that in the Util4 “frequency standard” menu you have option 4 selected, “internal 10MHz out”,
There are also two versions of the main instrument planar, which one is being used a AA1 or a AA1/1.
At risk of repeating myself, the main problem with these signal generators is the surface mount electrolytic capacitors failing, the smaller the physical diameter of them the more likely that they are to have failed. The smaller diameter ones appear to have problems with the electrolyte seals failing.
73 George G6HIG in Dover UK.