Hey forum,
i recently got a boat anchor called FSA Spectrum analyzer by Rohde & Schwarz.
I bought it with the both analyzer bus cables missing, successfully built the harness myself and was happy to have got a probably fully working unit.
Now some weeks later I get constant error messages regarding level error of the 2nd local oscillator.
I want to check the single module without the whole unit on the desk but I have no idea where to start checking the circuitry.
The signal STL2LO on Sheet 2, section F7, is staggering around by times. Some times is about 120-140mV, the next moment its around 300mV. A lever lower than approx. 160mV triggers internal fault messages.
May I start to check the circuitry in the proximity of N2 / V43 on sheet 2 ?
Seems to me like the feedback of the output amp.
X4 on sheet 1 is the 200MHz reference input
X7 on sheet 2 is the 2.2GHz / 18dBm LO-output
hack on,
induzer  :-/O
    • Laborlogbuch
Re: R&S FSA - Issues with level glitches of 2nd LO
« Reply #1 on: August 19, 2019, 07:27:01 am »
so I got the problems solved by doing a full alignment of the 2nd LO Module.
Offline loulou31
« Reply #2 on: September 29, 2021, 10:27:10 am »
Hi
I bought a FSAS spectrum analyser and I would like to know how to have access to the service manual to this instrument.
It was in paper form, but I dont know if some has done some scan of it and could be available on CDROM or transfer file, even not free.
Thank you and best regards
Jean-Louis
Offline loulou31
« Reply #3 on: October 08, 2021, 01:18:17 pm »
I have found service manual of FSA ( not FSAS that includes tracking generator) on KO4BB site that it is very rich in manuals :
http://www.ko4bb.com/getsimple/index.php?id=manuals
Jean-Louis