Hi!
I have a Tektronix 2213A with a dead PSU that I'm trying to fix. After I got the scope, it worked fine for about 15 minutes. Then it emitted a bang/pooof and I could smell that something bad had happened... No visible smoke, though. I expected to see some black residue or something somewhere in the PSU or some other place in the scope, but there was no visible signs of anything bad. I have looked at every component. The caps look fine with no bulging, no burn marks. The fuse is still intact.
I removed the three RIFA caps, and they look fine and measure fine out of circuit. I will replace them later on, but for now I'm testing the scope without them. I guess they could have blown at any time, but I'm pretty sure they were not the problem. Could there be RIFA-type caps in the Schaffner filter in the metal casing on the IEC input plug? In that case I could understand that there was no visible trace of an explosion. The input plug is still working, though, so in that case some other components was damaged as a side effect, because the scope is not working now even though 230 VAC is coming into the scope.
I have check that the 230 VAC is reaching into the scope, past the power switch. I plugged the scope power cable into a power meter. When I connect the scope, the power meter reads 1 W. When I switch the scope on, it still reads 1 W...
When I measure the different voltage rails described in the service manual, I get 0 VDC on all of them, with no ripple (TP961, W968, W960, W956, W954). I have checked both with a DMM and another scope.
I then checked the PSU waveforms (I ran the other scope ungrounded for this, in order to use TP940 and TP950 as reference, but I was very careful when doing this). First I tried to use a two channel scope and use subtraction, but I was not confident that I was doing it right. I'm pretty sure I have done the measurements correctly now with the ungrounded scope).
The result was that waveforms 27, 28, 29, 30, 32, 33, 34, 35) were flat at 0V with no significant ripple.
I did a few measurements on the big metal encapsulated filter cap C906. It gets charged to 320 VDC with no ripple. The resistance across it is many MOhms. Measuring its capacitance gives 95 uF, which is fine.
The smaller cap C925 doesn't seem to get charged. I measure only some 100 mV on it (with and without Q9070 connected in the live circuit operation). Reistance across it is only around 100 Ohm (in circuit).
The DMM diode test on each of the rectifier diodes CR901-904 gives expected results with a forward voltage drop around 0.6 V in one direction and O.L. in the other.
I disconnected the connected W9077 which goes to the IRF710 N-channel MOSFET Q9070 on the heatsink. The FET looks fine visually. Could it "blow" without it being visible afterwards? I did some measurements on it following
http://electronicsbeliever.com/how-to-know-if-mosfet-is-defective/ . From the guide there it seems that the FET is defective:
Source-drain diode test on DMM: Positive probe on source and negative probe on drain: O.L. (should be 0.4V - 0.9V according to the web page above for N-channel MOSFETs). Reversing the DMM leads still gives O.L.
Source-drain resistance: O.L. for both polarities.
Source-gate resistance: O.L. for both polarities.
Drain-gate resistance: 94 Ohm for both measurement polarities.
These results look quite suspicious to me. Could this be the cuplrit even though I don't see any "burn marks" or other signs of any violent explosion on it?
I don't want to start replacing components willy nilly, so I'm hoping some of you guys could have a think about what is the most likely suspect.