I'm digging into the heart of the beast
I removed the Vertical Preamplifier board today to get to the hard-to-reach parts. All of the old resistors and capacitors were replaced with new. The old parts are so out of whack!
The only part yet to be replaced is the Selenium 100mA rectifier plate (SR321). I compared the Early 310 schematics with the newer 310A schematics. The Selenium rectifier plate was replaced by a Germanium diode (T12G, AKA 1N270) in the newer version 310A scope.
I found two NOS Tektronix T12G diodes (part# 152-0008-00) on Ebay and here they are
Now, within the 310 and 310A, there are only two resistors that were required to be a "matched" pair. They come off of the preamplifier board (R320 and R401). I put yellow dots on them.
So I decided to pull and measure them. They are marked as 1%, 1 MOhm resistors. I tested each one on my Fluke 45 multimeter. Each resistor was given one minute to settle down. One of the resistors measured 1.0144 MOhm and the other 1.0015 MOhm.
I purchased four of these Dale .1% CMF60 resistors and put them through the same test.
Two of these resistors settled at 1.0006 MOhm! They're going in
Here are the new Silicon rectifiers along with isolated turrets to mount the new 10 Ohm 2 Watt resistors. New PTFE 20 AWG wire is being used to connect the rectifiers to the rebuilt can capacitors. In the background, I bought PTFE 22 AWG striped wire to replace the old wires in the transformer area. I also used the spacers from the old Selenium rectifiers to lift the new rectifier mount off of the chassis.
With the hard-to-reach soldering done beforehand, I installed the rectifiers. I drilled out two of the four old rectifier mounting bolt holes and installed rubber grommets to pass the wires through. Next is to wire the transformer back up.
Most of the low-voltage section is now back in place with new wires attached. I wanted to repeat what I said about the wires that attach to the back of the power plug socket. I found one with a broken solder joint. It's because the plug is made out of heavy brass bar stock. Tektronix used very little solder. I cleaned the brass on the bench and pre-tinned the brass with ~740 degrees at the solder iron tip. The new joints will be rock-solid for another 50 years
After the large resistors were re-wired, the rebuilt capacitors were installed. The new Low-pass film capacitor (C661) was also installed just above the topmost rectifier too.
I'm over to the other half of the oscilloscope now, and working around C250 was no joke. Two soldering irons would've been needed to lift it from the ceramic strip without breaking it, so instead, I carefully removed the old Allen Bradley resistors from below and installed the new resistors the same way.
When I got to V150, a 6AL5 tube, I noticed pin 7 wasn't soldered to a lead tying both pins 2 & 7 (the plates of the tube) to capacitor C115. Maybe it had an intermittent problem? I don't know, but it's soldered-in now. C115 is a 5% 180 pf Mica cap, so I measured it. It came out to 200pf, a little high, so I purchased a new 1% Mica. One of the NE-2 neons nearby had rotted a leg off, so I replaced both neons with new ones. David Hess suggests using neons that have a little bit of radioactivity so that they have a consistent strike voltage whether in the darkness or the light in these circuits, so I did.
Here's the before and the after of this "quite-busy" Sweep Generator section of the scope
...and here's the rebuilt Vertical Amplifier portion. All pairs of resistors on the schematics are matched (under .2%). I really like these new Vishay resistors. Their smaller size fits these tight spaces much better than the AB's.
I rebuilt my Sencore Z-Meter last year. This is the most accurate capacitor it has tested so far! A Tektronix can (.01uf/.1uf/1.0uf)!
Now on to the cleaning, rebuilding, and lubricating of the switches. Can you see the split in that large tube capacitor
...and now the Time Base "trigger" switch is rebuilt and installed into the sea of new red resistors. I flushed and cleaned all switches with Deoxit F5 too. The contacts shine and the movements are smoooooth. The Time/Division switch is next with only just a few caps and resistors left to test/replace. I see the light