The heater power is "controlled" by the tip. The base station provides high power RF to the tip (at 13.56 MHz). This is absorbed by the tip, until it goes above the Curie temperature when its physical magnetic properties change, and it reflects most of the power back to the base.
I can see several possible causes of this issue:
-
Faulty passives in the handpiece (see below).
- Bad connection at the RF F-connector.
- Break or near-break of the conductor within the cable.
- Faulty power supply not able to maintain the >40 W RF power output.
Note that the handpiece is more than just a coax cable with a handle at the end, there is an inductor and capacitor near the tip connector. See
this thread and especially
this thread .
If possible, borrow another handpiece and/or power supply to test with. This will help narrow down the issue to the power brick or the handpiece.