Hey friends, got this nice Fluke 8060A and I was eager to place a new battery and start using it.

Since the 9V battery connector was corroded along with the battery, it broke when I pulled it out, so I was left with the raw red and black wires.

Hold a 9V battery onto the wires just to see if it powers on, and it kind of did but very shortly and digits seemed to not be complete, then it faded out.

Since I didn't had a good grip on the the small wires onto the battery tips, I have mistakenly reversed the polarity and the battery started to heat.

Now the multimeter does not power up at all and kind of puts short on the battery (measured about 2,3 Ohm on the batt wires).

Did I messed it up completely or there's some hope in repairing it? I have attached some pictures.

Thank you everyone for any inputs!
 

Offline stevopedia

  • Contributor
  • Posts: 14
  • Country: us
Re: My first serious multimeter (Fluke true RMS) and I messed it up. Help
« Reply #1 on: October 08, 2024, 06:45:39 pm »
There's a good chance this is an easy repair! Do you have access to a soldering iron and another multimeter?

If you look on page 139 the manual PDF (https://assets.fluke.com/manuals/8060a___imeng0300.pdf) you'll see the schematic for the main board of the meter. The part we're interested in, the power switch and such, is tucked away in the bottom-left corner. VR3 is a 12 V Zener meant to protect the meter from reverse voltage such as accidentally reversing the battery. In this case VR3 probably failed short, which means it did its job perfectly!

Turn on the 8060's power switch, and then with the other multimeter measure the resistance across the battery wires in both directions. If VR3 failed short, you should see a very low resistance in both directions. Page 135 of the PDF shows where VR3 is on the board: placed vertically and tucked under the display, right next to the battery eliminator jack. If you cut or unsolder the wire coming out of VR3 and the short on the battery input goes away, then you've found the cause. Remove the original VR3 from the board and replace it with a new 12 V Zener diode to restore the 8060's reverse polarity protection. The meter will work without VR3, but if you accidentally reverse the battery again and VR3 is missing you will cause irreparable damage to the meter.
« Last Edit: October 08, 2024, 06:47:30 pm by stevopedia »
 
The following users thanked this post: highcap

Online coromonadalix

  • Super Contributor
  • ***
  • Posts: 6589
  • Country: ca
Re: My first serious multimeter (Fluke true RMS) and I messed it up. Help
« Reply #2 on: October 08, 2024, 07:04:29 pm »
yep  vr3  could have shorted out to protect the main section
 
The following users thanked this post: highcap

Offline highcapTopic starter

  • Contributor
  • Posts: 24
  • Country: ro
Re: My first serious multimeter (Fluke true RMS) and I messed it up. Help
« Reply #3 on: October 08, 2024, 10:32:12 pm »
Thank you that's great news!

I will first replace that diode before doing anything else, considering how I messed up with the polarity.

I will post an update as soon as I have it solved.

Best.
 

Offline lowimpedance

  • Super Contributor
  • ***
  • Posts: 1264
  • Country: au
  • Watts in an ohm?
Re: My first serious multimeter (Fluke true RMS) and I messed it up. Help
« Reply #4 on: October 09, 2024, 02:09:04 am »
 And while your in there, check the miniature electrolytic caps for electrolyte leakage as this will damage the PCB.
Its a common issue with these vintage meters, search the forum for details of 'others' repairs and possible suggested replacement parts.
I would actually change them anyway as a preventative measure.
The odd multimeter or 2 or 3 or 4...or........can't remember !.
 

Offline Dave Wise

  • Regular Contributor
  • *
  • Posts: 164
  • Country: us
Re: My first serious multimeter (Fluke true RMS) and I messed it up. Help
« Reply #5 on: October 09, 2024, 06:12:25 pm »
When I got my 8060A a couple months ago it had probably sat unpowered for decades with daily and seasonal temperature swings but most likely in a dry location.  I decided to try it without preemptive electrolytic capacitor replacement, since I saw no evidence of fluid egress.

When I first powered it, it didn't work, displaying a few random LCD segments.  After about a half hour it began to work normally.  Later I removed the battery and probed the caps, with an ESR meter and an impedance bridge.  They all read acceptable ESR, C, and D.  I left them alone and it is still working.

Make of it what you will.

Dave Wise
 

Offline Xena E

  • Frequent Contributor
  • **
  • Posts: 443
  • Country: gb
Re: My first serious multimeter (Fluke true RMS) and I messed it up. Help
« Reply #6 on: October 09, 2024, 07:12:10 pm »
When I got my 8060A a couple months ago it had probably sat unpowered for decades with daily and seasonal temperature swings but most likely in a dry location.  I decided to try it without preemptive electrolytic capacitor replacement, since I saw no evidence of fluid egress.

When I first powered it, it didn't work, displaying a few random LCD segments.  After about a half hour it began to work normally.  Later I removed the battery and probed the caps, with an ESR meter and an impedance bridge.  They all read acceptable ESR, C, and D.  I left them alone and it is still working.

Make of it what you will.

Dave Wise

Wise man indeed :)
Old good quality caps will reform just fine in most circumstances, blanket replacement without good cause is just inviting trouble.
 

Offline lowimpedance

  • Super Contributor
  • ***
  • Posts: 1264
  • Country: au
  • Watts in an ohm?
Re: My first serious multimeter (Fluke true RMS) and I messed it up. Help
« Reply #7 on: October 10, 2024, 12:28:34 am »
When I got my 8060A a couple months ago it had probably sat unpowered for decades with daily and seasonal temperature swings but most likely in a dry location.  I decided to try it without preemptive electrolytic capacitor replacement, since I saw no evidence of fluid egress.

When I first powered it, it didn't work, displaying a few random LCD segments.  After about a half hour it began to work normally.  Later I removed the battery and probed the caps, with an ESR meter and an impedance bridge.  They all read acceptable ESR, C, and D.  I left them alone and it is still working.

Make of it what you will.

Dave Wise

Wise man indeed :)
Old good quality caps will reform just fine in most circumstances, blanket replacement without good cause is just inviting trouble.

Sure by all means check your caps and make your own decision depending on what's its worth to you. BTW I am the same, I don't replace caps blindly, but some equipment are notorious for their leaking caps and the resulting damage, that could have been avoided . For example, Keithley 2000 series DMM's would you risk it !

Also I have an 8062a that lived all its life in an air conditioned lab and it had extensive damage from leaking electrolytic caps YMMV.

Above all though great to see another vintage meter back in action  :-+
The odd multimeter or 2 or 3 or 4...or........can't remember !.
 

Offline Xena E

  • Frequent Contributor
  • **
  • Posts: 443
  • Country: gb
Re: My first serious multimeter (Fluke true RMS) and I messed it up. Help
« Reply #8 on: October 10, 2024, 11:34:08 pm »
When I got my 8060A a couple months ago it had probably sat unpowered for decades with daily and seasonal temperature swings but most likely in a dry location.  I decided to try it without preemptive electrolytic capacitor replacement, since I saw no evidence of fluid egress.

When I first powered it, it didn't work, displaying a few random LCD segments.  After about a half hour it began to work normally.  Later I removed the battery and probed the caps, with an ESR meter and an impedance bridge.  They all read acceptable ESR, C, and D.  I left them alone and it is still working.

Make of it what you will.

Dave Wise

Wise man indeed :)
Old good quality caps will reform just fine in most circumstances, blanket replacement without good cause is just inviting trouble.

Sure by all means check your caps and make your own decision depending on what's its worth to you. BTW I am the same, I don't replace caps blindly, but some equipment are notorious for their leaking caps and the resulting damage, that could have been avoided . For example, Keithley 2000 series DMM's would you risk it !

Also I have an 8062a that lived all its life in an air conditioned lab and it had extensive damage from leaking electrolytic caps YMMV.

Above all though great to see another vintage meter back in action  :-+

Yeah, that's absolutely right, there are some brands of capacitors that are just awful, replace on sight junk. Ever come up against the red/black molded case Plessey electrolytic capacitors from the 1960s? Haven't found a good one yet, it's a case of knowing what's normally OK, (perhaps most brands), and what is garbage.

It's just that there are those people that just dive in and shotgun the job thinking that just replacing a bunch of capacitors will fix any fault. The next type of personality just do it to good working equipment, because they think that's what you do for a restoration.... I've replaced all the caps in my xxxx and now it won't work, is a common comment on the help forums.
 

Offline Dave Wise

  • Regular Contributor
  • *
  • Posts: 164
  • Country: us
All of my 8060A caps are Chemi-Con, unmarked but probably SR/SRC/SRJ series except for the RMS coupler which is LL series.  All 85C temp rating and probably 1000h service life - the latter of which has been exceeded by orders of magnitude, but sometimes you get lucky - I run Tektronix oscilloscopes from the 1950's with mostly original electrolytics.  OTOH I do check them now and then...
 