Hello
I already have a working E3633A. He had a damaged transistor Q5.
The power supply is functional but has an offset in the measurement of current, which I cannot calibrate.
After switching the output on, the display shows 13mA.
However, the output current is 0mA because there is no load.
I suspected that this was an error generated by the ampmeter in the power supply.
To remove this incorrect indication I even gave a 20k potentiometer to compensate input offset on the U44 opamp. I connected 20k pot between between pins 1 and 8 (Vos trim) and 7 – it is responsible for measuring the current on the display.
Thanks to this, I could set 0mA on the display when measuring current.
However, CC settings were still incorrect.
To align the CC settings and current measurement, I performed the calibration procedure again.
To my surprise, after calibration the previous 13mA result returned on the display, the potentiometer settings were leveled.
So calibration “works” because it removes the offset caused by the potentiometer, which appears on U44.
Potentiometer re-adjustments 0mA on display and calibrations procedure caused the same behavior, so I removed the potentiometer.
The -13mA current is generated from the current source R147 and Q7.
When the CC setting is below 13mA, the output current begins to flow in the opposite direction from the Q7 current source and the voltage is negative -0.4V .
When setting CC is 0.001mA the current flowing to the output is -12mA.
CC= 0mA output curent is -13mA
CC=13mA output curent is 0mA
CC= 20mA output curent is 8mA
Maybe someone knows where the problem is.
Michal