Author Topic: Keithley 2110  (Read 545 times)
Offline hitecelecTopic starter
  • Country: in
Keithley 2110
« on: September 27, 2024, 09:09:54 am »
hello Folks
Keithley 5 1/2 digit dmm having a problem
any hint or schematics available?
thanks in advance
Online Kleinstein
Re: Keithley 2110
« Reply #1 on: September 27, 2024, 09:28:14 am »
The Keithley 2110 is supposedly the same or very similar to picotest 3510/3511. There may be some service manual available from that side. However don't expect schematics.
The following users thanked this post: hitecelec
Online squadchannel
  • Country: jp
  • deepl translate user
Re: Keithley 2110
« Reply #2 on: September 27, 2024, 10:03:00 am »
2110 is M3510A. There is no service manual available.
2100 = M3500A, is available. It may be helpful.
Online Kleinstein
Re: Keithley 2110
« Reply #3 on: September 27, 2024, 02:39:52 pm »
The M3500A service manual is of very limited use. That meter is closely related to the K2000. The M3510 / K2110 are build around a SD ADC chip and because of the lower voltage range of the ADC also a different front end and amplification.
What are the problems of the meter ? Depending on the faults one may be able to narrow the problem down to a part of the meter.
Some faults can also be detected from a careful optical inspection of the suspect area.
If nothing helps a crude reverse engeniering of the relevant part could be possible - usually there are not that many custom parts.
  • Country: in
Re: Keithley 2110
« Reply #4 on: September 28, 2024, 08:07:50 am »
Hello
thanks for the feedback. Unit doesn't respond volts/Amp/ Ohms Inputs
PFA video
thanks in advance
« Last Edit: September 28, 2024, 08:36:54 am by hitecelec »
Offline J-R
Re: Keithley 2110
« Reply #5 on: September 28, 2024, 09:06:12 am »
Did you check the basics, such as various supply voltages on the PCB?
Block diagram and some other analysis here: https://gerrysweeney.com/keithley-2110-dmm-teardown/
  • Country: in
Re: Keithley 2110
« Reply #6 on: September 28, 2024, 09:21:22 am »
Will check and revert
Offline tooki
Re: Keithley 2110
« Reply #7 on: September 28, 2024, 10:10:53 am »
You sure about that? The K2000 is a "true" Keithley design. I think you're thinking of the K2100.
Online Kleinstein
Re: Keithley 2110
« Reply #8 on: September 28, 2024, 12:24:02 pm »
AFAIK the Keithley2100 and PicotestM3500 are essentially identical. More like Keithley design (at least to a large part) and maybe manufactured by picotest.
Offline tooki
Re: Keithley 2110
« Reply #9 on: September 28, 2024, 12:52:02 pm »
Exactly. But above, you wrote that the M3500A is "closely related to the K2000", not the K2100. We have no reason to believe the K2000 is related in any way.
Online Kleinstein
Re: Keithley 2110
« Reply #10 on: September 28, 2024, 01:52:57 pm »
There is a repair thread on the K2100. Not identical to the K2000, but still very close, up to the point of having corresponding part lables for many parts of the front end. From the pictures I saw, the anlog part looks like an only slightly modernized K2000.
Anyway the K2110 is a very different design.
The error description is still not specific. At least it seems the display still works and not error codes given.
Is there still at least some noise ? Is it still reacting to keys and switching a relay or when switching current ranges or from the 1 V to 1000 V range ?
It the temperature reading still working ?
Are the ohms ranges (e.g. 2W 100 Ohm should give some 1 mA) still sending out some current (e.g. measure with a different DMM) ?
Offline tooki
Re: Keithley 2110
« Reply #11 on: September 28, 2024, 03:57:00 pm »
Really?!? I'll have to look into that! Thanks for the tip.