@bdivi - Yes, you can calibrate using RS232 and specify the actual voltage, e.g. :CAL:PROT:DC:STEP3 10.00275. I can only dream of a 732 and my only externally verified ref is a Hao Qi Xin with 10.00275. My Keithley 2000 was bang on with this reference but my 2015 was out 250uV or so - but providing the actual voltage over SCPI cal and both meters now agree with each other.
A while back I got hold of a nice ESI RV622A Dekavider fairly cheap which is good for 700V and so should be able to calibrate the 100V as well as the oddball 19V ranges etc on my other meters too. I just need a stable HV supply.