Hello good people of the EEVblog Forum
To me: I live in Switzerland and am a soon to be student that studies electronics engineering.
To the subject: A while back I bought a defective Keithley 2002 from an auction site and put it aside. Now I have some time in hand and would like get it running again.
I have read through the excellent post
https://xdevs.com/fix/kei2002_u2/ and some other posts on the forum that successfully repaired a Keithley 2002 or similar.
To the device: It came with the cap C117 and rectifier CR113 missing. To be seen at 'pic2'. The LCD lights up fine but on whatever you choose, Voltage, Ampere etc. it shows 'overflow', as to be seen on 'pic1'. I have measured the Voltages and they seem as following:
5V -> measured about 4.1V
+15V -> measured about 2.2V
-15V -> measured about 1.1V
The device fails so many tests, that I did not bother to look into those.
After that I carefully removed all the electrolytic caps on the analog board. The reason for that is that, the inside really smelled fishy and I have heard many times that it is good practice. 'pic3'
To no ones surprise it was a great idea because, some of the caps already started leaking. 'pic4 and 5'
The caps I removed look fine, besides one 'pic6 and 7' The one on 'pic7' is not near its rated capacitance, it measures some nanofarads. I hope that that is why the Keithley did not work
I cleaned everything nicely with isopropanol and looked at the traces. None of them see to be damaged severly, as far as I can see no damages have been done. 'pic8'.
I am currently making a basket with all the caps I need (electrolyte and tantal). I will install everything this friday probably and here is where I need your help.
I am suspecting that the device wont work, because that would be too easy woudlnt it be?
If anyone has any other suggestions or places that I should check while waiting for the parts, Id be really greatful to hear them.
Thanks everyone in advance and greetings
SelectLOL
Sorry for the image quality, had to resize alot to be able to post