Hi HighVoltage,
it is a high sensative null-meter (1µV, 10x Scale).
Thus I would highly interested in a top view picture of the mounted tool (but with open lid). I am interested into the shielding construction in general.
I assume it is battery powered only - correct?
I did restauration of two a not that sensitive (3µV) HP149A, which could powered also from AC. Hee I found the internal wiring (where are the wires going aside) had some major effect in noise sensivity.
PeludeI will take those pictures tomorrow for you ...
Yes, my meter is battery powered only!
But ... I found the 230V PSU part of it separately and just bought it. That one is dead and needs repair as well.
My 155 has been on a dual power supply and turned ON for about 10 days now.
I did a first calibration 10 days ago, based on the instructions in the Service Manual.
All the POTS needed a few good movements before they got stable and reduced the noise.
It has applied 1.0000 V now and since then it seems the needle has not moved at all.
In the uV range it gets totally over sensitive. (That might be normal)
Getting the hands close to the instrument moved the needle right away.
It definitely needs a guard plugged in and probably some quality and shielded cables.
It is a surprise, that this instrument does not have a triaxial connector, as the Keithley Electrometer have, for instance.
dr.dieselWhat is wrong with yours ?
I am glad we have the full schematics and probably can find all problems.