Hi there,
I have a 1999-2000 Unit that is approaching the end of its current calibration interval and I took note of the "wet" Capacitors inside:
Electrolytic Capacitors:
A1:
A2:
A3:
A4:
C7,C8: 2200µF 35V 0180-4033 16,5mm dia. 32mm long (1189-2253-ND)(DK)
C9 15kµF 16V 0180-4034 36mm dia. 26mm long (493-8615-ND)(DK)
C10-15: 1µF 35-50V Hermetic
C17: 22µF 50V 0180-3913 8mm dia. 16mm long (1189-50YXM22MEFRT15X11CT-ND)(DK)
A5:
A6:
C1,C11 680µF 50V 0180-4032 16mm dia. 21mm long (493-1914-ND)(DK)
C200-01 47µF 63V 0180-3828 8mm dia. 12mm long (P19619CT-ND)(DK)
C8 8300µF 35V 0180-3898 25mm dia. 67mm long axial (493-8635-ND)(DK)
C202 330µF 25V 0180-3830 10mm dia. 12mm long (1189-1891-ND)(DK)
C203 220µF 16V 0180-3829 8mm dia. 12mm long (493-14723-1-ND)(DK) or (1189-4157-1-ND)(DK)
A7:
A9:
Tantalum Capacitors
A1:
C3: 3,3µF 15V 0180-0210 Hermetic
C17: 1µF 35V 0180-0291 Hermetic
C20/21: 2,2µF 20V 0180-0197 Hermetic
A2:
C211: 15µF 20V 0180-1746 Hermetic (399-11291-ND)(DK)
C701-706 " Hermetic
C709-710 " 6,5mm dia. 12,5mm long (1189-50YXM22MEFRT15X11CT-ND)(DK)
all caps except C702 and C704 are hermetic, those are 22µF 50V
Al-Electrolytic Nichicon
A3:
C146/47 15µF 20V 0180-1746 Hermetic
C181 33µF 10V 0180-0229 Hermetic
C303 "
C353 "
C250/51 4,7µF 10V 0180-0309 Hermetic
C252/53 4,7µF 35V 0180-0100 Hermetic
A4:
A5:
C150-55 1µF 35V 0180-0291 Hermetic
A6:
C2 6,8µF 35V 0180-0116 Hermetic
C4 2,2µF 20V 0180-0197 Hermetic
C105 "
C304 "
C12 4,7µF 35V 0180-0100 Hermetic
C406/07 1µF 35V 0180-0291 Hermetic
A7:
A9:
I will add the digikey or mouser part numbers for replacement caps soon.
One more thing, does anyone know the lead spacing on A4C9, I forgot to measure that...
Cheers,
Ole
Edit: This is list is not yet finished, so if you think that a capacitor doesnt fit please let me know