Fluke 8060A switch trouble
« on: August 05, 2014, 07:46:24 pm »
I have a Fluke 8060A multimeter that needs some help. Hopefully Dr. Taylor can shed some light. It's a recent eBay purchase described as "won't power up." Cosmetically, it was very clean and inside I easily found several old electrolytics that had "given up the ghost." At this point, the meter powers up fine, passes the power-on self test and appears to be working. Yea!

Not so fast. I've found that the range push-buttons are not being recognized correctly. Specifically, on performing the "switch decoding self-test" described on p. 5-22 of the manual, the 200 uA range switch and the 2000 mA range switch are both recognized as the 200 mA range setting. The 2 mA and 20 mA ranges decode correctly. I've tried cleaning the switches with De-Oxit without luck. I've also removed and re-seated the main measurement acquisition chip. I'd be grateful for any ideas.
 

Offline retiredcaps

  • Super Contributor
  • ***
  • Posts: 3575
  • Country: ca
Re: Fluke 8060A switch trouble
« Reply #1 on: August 05, 2014, 08:04:31 pm »
I don't know if this will help or not, but usually when the caps on the 8060A go bad, they spew out their electrolyte from the bottom onto the pcb and can wick their way up onto other components.  A good IPA cleaning with a toothbrush and thorough drying might help?

PS. Did you clean main IC with IPA?

See

http://mrmodemhead.com/blog/fluke-8060a-repair/
 

Offline vindolineTopic starter

  • Supporter
  • ****
  • Posts: 328
  • Country: us
Re: Fluke 8060A switch trouble
« Reply #2 on: August 06, 2014, 02:21:12 am »
I don't know if this will help or not, but usually when the caps on the 8060A go bad, they spew out their electrolyte from the bottom onto the pcb and can wick their way up onto other components.  A good IPA cleaning with a toothbrush and thorough drying might help?

PS. Did you clean main IC with IPA?

See

http://mrmodemhead.com/blog/fluke-8060a-repair/

I cleaned it pretty well with IPA when I replaced the caps. I'll give it a second bath and see if that clears it up.
 

Offline lowimpedance

  • Super Contributor
  • ***
  • Posts: 1264
  • Country: au
  • Watts in an ohm?
Re: Fluke 8060A switch trouble
« Reply #3 on: August 06, 2014, 03:31:42 am »
Not sure if this helps, but DrTaylor did reply to a question kind of similar in the 'old Fluke multimeter' thread, see below top post on P12.

https://www.eevblog.com/forum/testgear/old-fluke-multimeters/165/
The odd multimeter or 2 or 3 or 4...or........can't remember !.
 

Offline vindolineTopic starter

  • Supporter
  • ****
  • Posts: 328
  • Country: us
Re: Fluke 8060A switch trouble
« Reply #4 on: August 06, 2014, 03:26:30 pm »
Not sure if this helps, but DrTaylor did reply to a question kind of similar in the 'old Fluke multimeter' thread, see below top post on P12.

https://www.eevblog.com/forum/testgear/old-fluke-multimeters/165/

Thanks for the suggestions and pointing me to Dr. Taylor's earlier post. Unfortunately, I've narrowed down the trouble to mechanical failure in one or more of the switches. I'm afraid this unit may be a parts donor now  :(
 

Offline Excavatoree

  • Frequent Contributor
  • **
  • Posts: 906
  • Country: us
Re: Fluke 8060A switch trouble
« Reply #5 on: August 06, 2014, 04:55:44 pm »
Not sure if this helps, but DrTaylor did reply to a question kind of similar in the 'old Fluke multimeter' thread, see below top post on P12.

https://www.eevblog.com/forum/testgear/old-fluke-multimeters/165/

Thanks for the suggestions and pointing me to Dr. Taylor's earlier post. Unfortunately, I've narrowed down the trouble to mechanical failure in one or more of the switches. I'm afraid this unit may be a parts donor now  :(

I might have an 8060 with a corroded main board (too many years with capacitor goo on it) with a good switch assy.  (well, I suppose I don't know for sure) If I have one, you're welcome to it if you want to try changing the switch assy.  Let me know if you're interested and I'll dig it up.

I know Retiredcaps and Modemhead repair these boards, but once the get as far gone as a couple of mine are, I don't see the point.  Especially when I don't have enough LCDs to go around.  (or buttons, LOL)

 

Offline vindolineTopic starter

  • Supporter
  • ****
  • Posts: 328
  • Country: us
Re: Fluke 8060A switch trouble
« Reply #6 on: August 06, 2014, 11:18:22 pm »
Not sure if this helps, but DrTaylor did reply to a question kind of similar in the 'old Fluke multimeter' thread, see below top post on P12.

https://www.eevblog.com/forum/testgear/old-fluke-multimeters/165/

Thanks for the suggestions and pointing me to Dr. Taylor's earlier post. Unfortunately, I've narrowed down the trouble to mechanical failure in one or more of the switches. I'm afraid this unit may be a parts donor now  :(

I might have an 8060 with a corroded main board (too many years with capacitor goo on it) with a good switch assy.  (well, I suppose I don't know for sure) If I have one, you're welcome to it if you want to try changing the switch assy.  Let me know if you're interested and I'll dig it up.

I know Retiredcaps and Modemhead repair these boards, but once the get as far gone as a couple of mine are, I don't see the point.  Especially when I don't have enough LCDs to go around.  (or buttons, LOL)

That's really generous of you. If you can find the damaged unit, I'll give the transplant a try (nothing to loose!). Send me a pm and I'll give you my contact info.
 