Author Topic: HP 54504 A - Gain Chan 1 Calibration error  (Read 824 times)
Offline DiscipleTopic starter
« on: January 25, 2021, 10:01:37 am »
Greetings Gentlemen,
I have an issue with my HP54504A. When I do the self calibration operation for channel 1 I get the error : Failed Gain Cal Chan 1 0000 0000 0010 0000. All other calibration test are passed with success. The calibration is passed with sucess for channel 2.
When I go to the service menu, I have for Chan 1 gain the 0020H error code.
I search a clue on the service manual but I found nothing. Do you have any idea where I can start to search ?
Thank a lot for your help,
Disciple
« Last Edit: July 06, 2021, 10:11:42 am by Disciple »
Online Runco990
Re: HP 54504 A - Gain Chan 1 Calibration error
« Reply #1 on: January 25, 2021, 07:45:24 pm »
The input attenuator may be damaged or just dirty.  Try cleaning it or swap around to see if the problem moves.  You DON'T want it to be a custom IC.  There is some talk about the input attenuators, in particular the 50ohm switch.  Swap them first, PITA I know.
Offline SoundTech-LG
« Reply #2 on: January 27, 2021, 07:29:52 pm »
Set Chan 1 50 ohm input impedance to ON. then, measure the input resistance. If higher than 50 ohms, likely the relays are problematic, possibly a damaged or open 50 ohm resistor... Sometimes switching them from 1meg to 50 ohms repeatedly may temporarily clear a relay problem. Then try a reCal to see what happens.
Online Runco990
« Reply #3 on: January 28, 2021, 12:36:00 am »
That's good advice.  And simple.  The relays don't get switched much.  A Tektronix tech told me they take the problem attenuators and "buzz" them for an hour to self clean the contacts.
However, they had a special setup for that.
On the HP scope... yep... try manually switching it if possible.  Otherwise you have to pull the attenuators and manually clean them.
« Reply #4 on: July 06, 2021, 10:14:52 am »
Thank you all for your help. I was offline during few month due to a lot of work. I will try during my holiday to finish the repair.
Thank a lot again !