HP 3465A DMM - zero ohm problem...
« on: April 03, 2016, 08:10:15 am »
My HP 3465A has a problem to display zero in the lower ohm ranges (10, 100, 1k and 10k).
The offset is about 1-2 ohms. At the beginning it was up to 10 ohms, but changing on moving the gang switches.
So I tried to clean the switch contacts with a few drops IPA per switch. After moving each switch a few times the offset reduced to 0.3 ohms.
I also cleaned the 4mm input sockets using IPA and Q-tips. The first result was not too bad: 0.17 ohm.
But over night the problem came back (1- 2.5 ohm offset).  :-\
BTW: what's the specification on shorted input?

What can I do now?
I don't want to contaminate the whole gang switch area with IPA residue by using more than a few drops directly to the contacts.
I also dont' want to remove the lubrication of the contacts completely...



(Edit: picture borrowed from IntelJoe)
 
« Last Edit: April 03, 2016, 10:17:40 am by e-doc »
 

Online tautech

  • Super Contributor
  • ***
  • Posts: 29482
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: HP 3465A - zero ohm problem...
« Reply #1 on: April 03, 2016, 08:55:38 am »
Often these sort of problems in CRO's are remedied with switch "exercise".
Tarnished contacts result in "noisy" changes of range selection and the contact lubrication is best totally removed with either IPA or as many recommend Deoxit along with exercise.

Of course the contacts are gold plated and they need be protected from wear with an appropriate lubricant.
Avid Rabid Hobbyist.
Some stuff seen @ Siglent HQ cannot be shared.
 

Offline e-docTopic starter

  • Contributor
  • Posts: 32
  • Country: de
Re: HP 3465A - zero ohm problem...
« Reply #2 on: April 03, 2016, 09:46:13 am »
I forgot to say it is a DMM not an oscilloscope, but you are right, the contacts often causes problems...
I think DMM switches are more problematic because isolation usually must be much better and contact resistance usually must be lower than in oscillosope switches.

What is the best (easiest) way to remove the lubricant and clean the contacts without complete disassembling the switches or instrument?
Is using a few drops of cleaner and moving the switch enough? I just did it this way with not too much success...

How to "relubricate"?
 
« Last Edit: April 03, 2016, 09:59:51 am by e-doc »
 

Online tautech

  • Super Contributor
  • ***
  • Posts: 29482
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: HP 3465A DMM - zero ohm problem...
« Reply #3 on: April 03, 2016, 10:06:11 am »
See the posts in this thread, there should be all the info you need:
https://www.eevblog.com/forum/repair/vintageclassic-renovation-techniques/
Avid Rabid Hobbyist.
Some stuff seen @ Siglent HQ cannot be shared.
 

Offline e-docTopic starter

  • Contributor
  • Posts: 32
  • Country: de
Re: HP 3465A DMM - zero ohm problem...
« Reply #4 on: April 03, 2016, 10:20:33 am »
Thanks for the link.

Is "Deoxid" cleaner and lubricant worth the money?
Can I buy it in Germany?
 

Online tautech

  • Super Contributor
  • ***
  • Posts: 29482
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: HP 3465A DMM - zero ohm problem...
« Reply #5 on: April 03, 2016, 10:49:45 am »
Thanks for the link.

Is "Deoxid" cleaner and lubricant worth the money?
Can I buy it in Germany?
Don't know.

It is the most quoted contact cleaner on the forum AFAIK.

Here's the manufacturers page:
http://store.caig.com/s.nl/sc.2/category.188/.f
Avid Rabid Hobbyist.
Some stuff seen @ Siglent HQ cannot be shared.
 

Online The Soulman

  • Super Contributor
  • ***
  • Posts: 1037
  • Country: nl
  • The sky is the limit!
Re: HP 3465A DMM - zero ohm problem...
« Reply #6 on: April 03, 2016, 11:48:13 pm »
How do you short the inputs? 0,17 ohms doesn't sound that wrong to me for a 2 wire ohms meter.
 

Online Vgkid

  • Super Contributor
  • ***
  • Posts: 2727
  • Country: us
Re: HP 3465A DMM - zero ohm problem...
« Reply #7 on: April 04, 2016, 01:34:24 am »
That is fine for a 2 wire measurement. Might even be acceptable for some 4-wire multimeters.
If you own any North Hills Electronics gear, message me. L&N Fan
 

Offline e-docTopic starter

  • Contributor
  • Posts: 32
  • Country: de
Re: HP 3465A DMM - zero ohm problem...
« Reply #8 on: April 04, 2016, 12:02:34 pm »
How do you short the inputs? 0,17 ohms doesn't sound that wrong to me for a 2 wire ohms meter.
I use one of these:


0.17 ohms is OK, but now I have up to 80 ohms (fluctuating).
 

Offline babysitter

  • Frequent Contributor
  • **
  • Posts: 899
  • Country: de
  • pushing silicon at work
Re: HP 3465A DMM - zero ohm problem...
« Reply #9 on: April 04, 2016, 03:33:27 pm »
Teslanol Oszillin T6 bekommst Du recht leicht... Finger weg von CRC Kontakt 60 ohne Nachbehandlung, das Zeug macht zwar schön sauber, aber nach dem reinigen versucht es weiterzureinigen, greift also Zeug an. Mit einer Nachbehandlung wie ausspülen aber verwendbar.
I'm not a feature, I'm a bug! ARC DG3HDA
 

Offline mzacharias

  • Frequent Contributor
  • **
  • Posts: 710
  • Country: us
Re: HP 3465A DMM - zero ohm problem...
« Reply #10 on: April 04, 2016, 03:56:03 pm »
Any resistance in the ohms "path" will cause this. I had an issue with my Fluke 8840A where (apparently) dissimilar metals may have caused this where the front input jacks attach at their rear side. The small metal plugs which attached to the rear of the jacks appeared to have a "zinc-oxide" powdery appearance. Cleaning them, tightening and reconnecting fixed the issue. Working flawlessly for three years plus since then. Don't rule out any relay, switch, etc in that path. The fact that it is somewhat intermittent works in your favor. You can push, poke, tap around the area to help localize the problem. Using a 'scope will help you "see" any noisy connections.
 

Offline e-docTopic starter

  • Contributor
  • Posts: 32
  • Country: de
Re: HP 3465A DMM - zero ohm problem...
« Reply #11 on: April 04, 2016, 04:53:57 pm »
Teslanol Oszillin T6 bekommst Du recht leicht... Finger weg von CRC Kontakt 60 ohne Nachbehandlung, das Zeug macht zwar schön sauber, aber nach dem reinigen versucht es weiterzureinigen, greift also Zeug an. Mit einer Nachbehandlung wie ausspülen aber verwendbar.
Danke!
400ml only 5€  :)
https://www.buerklin.com/de/katalog/Kontakt-und-Tunersprays-Oszillin-Typ-Teslanol-t6-L636100.html

Hört sich gut an:
Quote
keine Frequenzveränderungen im VHF/UHF-Bereich, Übergangswiderstände werden schnell abgebaut, imprägnierend gegen Feuchtigkeit und Verschmutzungen, besonders ergiebig, mit Microgleitfilm, der nachhaltig die Kontakte konserviert und damit länger funktionsfähig macht. Keine Nachbehandlung notwendig, reinigt, pflegt und schützt in einem Arbeitsgang auch hochwertige Kontaktelemente und Steckverbindungen. Der Abrieb dünner Metallschichten wird reduziert und beschädigte Edelmetallschichten werden vor Korrosion geschützt.
Hoffentlich Tatsache und kein Voodoo...

"Kontakt 61/60"
Eines davon war aggressiv, hat sich nach ein paar Jahren durch die Spraydose "gefressen".
Weiß bloß nicht mehr ob es das "60" oder das "61" war. Habe seitdem lieber die Finger davon gelassen (von beiden).

PS: English translation:
"Kontakt 60/61"
One of these two was so aggressive, the can was eaten through from the inside after a few years of storage.
Don't know if it was "60" or "61". Kept my hands off it since.



« Last Edit: April 05, 2016, 12:40:48 pm by e-doc »
 

Offline babysitter

  • Frequent Contributor
  • **
  • Posts: 899
  • Country: de
  • pushing silicon at work
Re: HP 3465A DMM - zero ohm problem...
« Reply #12 on: April 04, 2016, 07:06:54 pm »
Habe mich schon seit 13 Jahren nicht mehr so recht drum gekümmert: Das Aggresive ist das 60, ist auch ein bischen legitim für ein Reinigungsmittel. 61 ist ein öliges, könnte konservieren. Kontakt WL war meines Wissens das richtige um das 60 nach Einsatz wegzumachen. Habe das Teslanol schon eine Weile und bin dabei geblieben.



I'm not a feature, I'm a bug! ARC DG3HDA
 

Online The Soulman

  • Super Contributor
  • ***
  • Posts: 1037
  • Country: nl
  • The sky is the limit!
Re: HP 3465A DMM - zero ohm problem...
« Reply #13 on: April 04, 2016, 07:31:45 pm »
Allo, bitte auf englisch weiter gehen oder anderes sollten die andere neugierig werden!  :-DD

Did you resolder all the switches and contacts jet? It could easily be a bad solder joint..
 

Offline e-docTopic starter

  • Contributor
  • Posts: 32
  • Country: de
Re: HP 3465A DMM - zero ohm problem...
« Reply #14 on: April 04, 2016, 08:36:09 pm »
Allo, bitte auf englisch weiter gehen oder anderes sollten die andere neugierig werden!  :-DD
Yes, Sir! ;)

Did you resolder all the switches and contacts jet? It could easily be a bad solder joint..

No, I did not resolder yet. I first have more to disassemble...