Author Topic: Tektronix 2465 Calibrator Broken  (Read 539 times)
Offline mawyattTopic starter
« on: May 07, 2020, 03:46:08 pm »
Hello,
New here and thought about posting a success repair for restoring the Calibration function. I purchased a couple 2465 off eBay, each had their problems. I've replaced many of the usual power supply capacitors (waiting for a few others) as indicated by many others, but the Calibrator waveform is not correct as you can see. I traced this out and after a little circuit analysis thought the cause was a defective zener diode, this is a special Motorola part. I replaced the Zener with a 1N4733A zener and this fixed the Calibrator waveform.
Best,
Mike
« Last Edit: May 07, 2020, 03:48:16 pm by mawyatt »
~Wyatt Labs by Mike~
The following users thanked this post: alm