Hi all, I recently acquired an Advantest TR6143 SMU from Japan for cheap on the 'bay. It has six banana binding posts on the front and three triax connectors on the back.
Some quick specs from the datasheet:- Sources up to 110V (0.5A max) and up to 2A (32V max);
- Source (and measure) resolution of 10μV (320mV range) and 1nA (32μA range);
- Stable as a voltage source into 1μF for all current ranges in SLOW mode.
I find it quite interesting that the unit is so tolerant of capacitive loads (relative to its western counterparts such as e.g. Keithley, HP/Agilent), even in FAST mode (100nF from the 3.2mA range upward), moreso considering that it's fairly old. I found a
bulletin from 1992 so it has to be at least that old. If I'm not mistaken, that dates it around the time that Keithley released their 236/237/238 SMUs, which only tolerate up to 20nF across the board.
I'm not sure when exactly Keithley and Agilent started adding a "high capacitance mode" to their SMUs, perhaps someone else here knows. The relatively new
Keithley 2450 for example still supports up only to 20nF in normal mode, and up to 50μF in "High-C mode". As far as I can remember, this mode roughly boils down to lowering the gain bandwidth of the integrator by a factor 100 or so, and increasing the time constant of the current sense resistors by another factor 25 or so. Those rough numbers can be found from the response times cited in the datasheets, I might be misremembering how the factor 2500 (50μF/20nF) is distributed. The concept is described in
this Keithley patent from 2007(!). Advantest must have done something similar in their TR6143.
Anyhow, I was hoping that I would be able to rewire it for European mains like I did for my
Advantest R6246. Unfortunately, no dice. The unit contains two (welded?-shut) black cans of quite different size as can be seen below (top left), with only two wires going into each. Judging by their size, the transformers experience unequal loading in operation so for that reason as well as some others wiring their input windings in series seems like a recipe for disaster. I guess I'll have to buy an external conversion transformer...
Since I had to rip the unit apart to find this out anyway, I took some pictures of the interesting stuff along the way. I'll update things as I get around to figuring out more about the unit, but no promises.