ok, so it seems i am only talking to/for myself in here but what the heck...I'll use it as my repair diary for now .
I measured the offset in baseline on all channels at different settings and the results are in the attached picture. The scope settings are also in there.
I appreciate if someone with a DPO4000 scope can share his/her results.Note that the unit did pass SPC once right after I upgraded the firmware but since that it has not been able to finish spc as it fails self test (just like before firmware upgrade) I think that one time was a fluke no pun intended!
Another thing that I noticed is that setting the input to GND does not completely kill the signal
it has a high pass behavior
For example on channel 1 (all channels behave
exactly the same) I tried a 1Vpp signal at 10KHz,10MHz,100MHz with input set to GND50
and GND_1M. Here are the results
@10KHz 0 div
@10MHz ~0.4 div (does not change with V/div)
@100MHz ~1.6div for GND50 and ~1div for GND_1M (no change with V/div)
can somebody with a DPO4000 share his/her measurements?Still I am not sure if any of these indicate the symptoms of the self test errors thoughas I mentioned, I found out that the
self test errors refer to U271/U370-B. Anybody has any idea what these chips are?
Is the demux or ADC or something else?
Also when the SPC failed, in :MFG:MOD 1 I can see a broken message in the error log saying this:
Ch3 linearity Cal failed max sig re-cente -->
it is cut off at the end of the line after centenormally you dont get any of these messages in the error log, just a number....