Today I received a package from Farnell with all the elcos and some diodes for the PSU of my HP 1715A (one of the old elcos on the 5V rail shorted and nicely charred a pair of rectifying diodes). After replacing all the caps and diodes (and an hour or so to reinsert the PSU board (all the wires from the transformer and some more are soldered in the PSU board)) I was finally able to turn the puppy on only to be met by the next problem.
CRT lights up (intensity dial works) but there is no trace. I tried setting it up like it's written in the service manual (all the dial positions and stuff) but nothing changed. Beam finder appears to do nothing although one time I saw a glimpse of a beam disappearing off the top of the CRT when I pressed it.
I have checked all the test points on the PSU and all the voltages (5, -15, 15, 20 (unregulated at 22.5), 53 and 115) are present, within spec and no ripple present.
On a side note... I started the troubleshooting procedure and I got to step 5 where my A8TP2 was 4.4V and A8TP7 was 4.7V (eg. none of the stated) which should indicate a problem with either Hold Off, Sweep length or Trigger. Is this connected or is it a separate issue?
I have to note that I don't have any signal applied to any of the channels but I believe I should still get a trace.
What I'm looking for is a starting point to find the source of the problem as I've never dealt with CRTs before.
Thank you and have a nice day