I would like to add this older thread, in case details are of use to anybody else, as I've just finished a similar repair on the HP141T spectrum analyser.
Referring to the CRT Safety Summary at front of "Operating and Service manual, HP141T Display Section",
the CRT and other parts I substituted are no longer available for HP141T as far as I could see on internet.
I think substituting parts in vintage hobby equipment is OK, but I would not recommend to anyone to do the following repair -of soldering new cables onto a cr tube. Alternatives are finding a donor HP141T . Also the4552B IF section has bnc outputs for vertical and scan that could be adapted for an external oscilloscope etc.
Another option might be to convert the 141T back to non-storage using either its existing tube without the 6600 V accelerator grid (see notes below) or a another available non standard tube.
With cosen's advice in #4 above I removed the tube (I think the original one from 1972) without pulling on the glass-to-plug adhesive.
After loosening the hose clamp, a small flat blade was used to pry the harness socket, going around the circumference unsticking the pins one by one "clicks" until they were all sliding, then the socket was pulled off by prying with a bigger screwdriver to avoid tension on the plug.
The black bezel on the front panel has to be removed by pinching vertically and rotating downwards.
Then after a few other details...., the tube can be pulled out from the front.
The failure here was also flashover of the 6600 V post accelerator -wire(2) to 4.2 volt storage mesh -wire(6) and 158 V to collector mesh -wire(602) inside the original ribbon cable running to the front electrodes.
The ribbon had delaminated and the copper tracks were corroded and green.
I replaced this cable with cores of AWG 22 in separate heatshrink sleeves running from the terminal, and soldered onto the electrode pins and insulated with hot glue. - covered the exposed tube with a heavy towel, wearing safety goggles and a mask.
Dielectric grease was added in the chassis connector so it squeezed out on re-clamping.
There were two other components failed on the boards:
a) 82 Volt regulator V1 on low voltage board ( Fig 8.10)
About a year ago the original neon regulator had failed and I converted V1 to a string of (5) 18 V 0.5 Watt zeners.
Now I am not sure if this failed ( one zener went open) and caused the 6600V to flashover in the delaminated cable or vice versa.
Anyway not liking the idea of using zeners string (essentially they are avalanche diodes) I have converted the regulator again, this time to a 3 terminal high voltage regulator TL783C using R1 = 68 Ohm and R2 = 2 series 2200 Ohm RN70, with a 0.68 Uf 630 V input capacitor.
b) diode CR26 on the Pulse Circuit Fig 8.16.
The diode failed to "leaky when reverse biased" with the 158 Volt on collector mesh -wire 602, resulting in the storage mesh being pulled up to about 120V and no trace.
The diode tested OK on a meter. This "leaky" failure of CR26 has also been reported on another forum.
I replaced it with a 1N4007 rather than a fast recovery diode.
Some notes that may assist in fault finding associated with the tube:
The HP141T here can be run in non storage mode as follows:
unplug and tie the yellow wire feeding -2350V to the tripler. That disables the tripler and the 6600 V supply to the crt
unlug and tie the yellow wire feeding -2450V to the intensity pot. Now the grid will float to cathode gun , -2444 V via CR28 etc
Press "Conv, non storage" on the front panel.
The HP141T will start up with a trace intensity not adjustable , but of sufficient brightness to be usable.
That gave me the incentive to continue poking around to pulse board to find the leaky CR26.
Here are some photos of tube removal etc
https://app.box.com/s/gqjeauodgacizrt4sxii2s9t3qu14quehttps://app.box.com/s/xbxtvfiyhkwj3u4css1e498hw6b1z5qzhttps://app.box.com/s/igdw42nl0pxzh71hnr8ox4wa4435ajhvhttps://app.box.com/s/vsv2wjahwe7j9vkaz4qlr1sajh9hxxo9https://app.box.com/s/e686ouc845a1qmjs4vooagvw2krszw6ohttps://app.box.com/s/7t3920pes7tzvfeql3jqa3h91s3jb36jPresently, The HP141T is all reassembled and fully working on non storage, slow and fast storage modes with a sharp spot and good brightness range.
https://app.box.com/s/b3mc6pu9l2ek4qn4soodq25qvln6m4irWishing good luck and patience to those keeping these old things working!