Go to PDF page 263 on the service manual here:
https://w140.com/tekwiki/images/d/d1/070-7672-00.pdf disconnect power supply output wires to completely isolate the PSU from the scope. If the 2245A uses PCB jumpers instead like traditional 22xx series scopes, desolder those. Measure voltages with a DMM and use a constant current load to measure the tripping current and make sure it matches up with the specs in the service manual.
If the PSU trips under loaded, trouble shoot the PSU. If the PSU works perfectly fine, trouble shoot the rest of the scope by disconnecting power to some parts of the scope and see what makes the PSU trip. Isolate the problem and troubleshoot the suspect board/board area.
Go to the Maintenance section of the manual and the sub section "Troubleshooting Techniques". Read that and then start your troubleshooting journey.