Hi All.
I have this Fluke 5502A calibrator, upon turn on, "Starting Up" Shows up and then it keeps cycling and rebooting to "Start Up". However leaving it be, it does boots into the calibrator sometimes and everything works perfectly, outputs are bang on.
To eliminate the problem, I tried removing all modules, leaving on the bare minimum of Processor, Front Display, and Power Supply. Its still the same. Power Supply test points are on the dot. So my suspect is the Processor A9 Module itself. Suspect is SRAM
I also noticed something weird, I went and checked the clock setting, it was not moving. Can someone help me to check on their 5500A/5502A/57xxA, if the clock setting is just setting feature without readback. I have checked all of Flukes 55xx service manual, they never mentioned anything about their RTC battery replacement. Anyone done this before?
Remarks:
57xxA stated on their service manual, its a simple replacement, but the architecture of 55xxA and 57xxA is different