Author Topic: fluke 343a -15V power failure for help  (Read 595 times)
Offline daisizhouTopic starter
  • Country: cn
« on: December 11, 2020, 12:08:37 pm »
A fluke 343a is being repaired and found that the -15V power supply has disappeared
Found that Q21-SM 4144 transistor is damaged.This is a transistor made by Motorola.Can't buy now
I want to know what should I do? Is there a replacement model
Thanks for help. thanks
1128926-0 [ Specified attachment is not available ] [ Specified attachment is not available ]
daisizhou#sina.com #=@
Offline wn1fju
Re: fluke 343a -15V power failure for help
« Reply #1 on: December 11, 2020, 04:58:08 pm »
The -15V is an auxiliary supply whose components do not seem terribly crucial.  The reference for the 343a, fortunately for you, is in the +15V supply.  I would think you could get by with just about any medium power general purpose PNP transistor for Q21.  Something like 50V C-E, a couple of hundred mA current and a 1 watt power dissipation.
  • Country: cn
« Reply #2 on: December 12, 2020, 01:54:04 am »
Thanks for the successful repair