Author Topic: Hameg HM5005 spectrum analyzer  (Read 3549 times)
Offline roddTopic starter
  • Country: br
« on: December 03, 2016, 03:09:41 pm »
I received one of these as a gift from a friend.
No trace at all so I am hopping this a power supply issue.
I searched the invoice net but could not find the schematics and/or the service manual.
Does anyone has a copy to share?
Thank you all in advance,
Roger
Enviado do meu iPhone usando Tapatalk
Online PA0PBZ
Re: Hameg HM5005 spectrum analyzer
« Reply #1 on: December 03, 2016, 04:58:53 pm »
I could not find anything free to download, but a few sites offer the document to download for $6-$10.
I found the service manual for the MH5010, but I have no idea how similar or different the power supply is.
http://www.ko4bb.com/manuals/92.110.177.228/Hameg_HM5010-11_Service_Manual.pdf
The following users thanked this post: kripton2035
Offline bktemp
« Reply #2 on: December 03, 2016, 05:39:20 pm »
Can you take a photo of the power supply board?
Most Hameg scopes/spectrum analysers when through several revisions. The early versions used a mains transformer + linear regulator. The later ones used a SMPS.
Often scopes and spectrum analysers used identical power supply boards, so even if you can't find a service manual for the spectrum analyser maybe an identical power supply board is used in a scope.
  • Country: br
« Reply #3 on: December 04, 2016, 04:04:54 pm »
I can't take a picture of it right now but it appears to be a SMS power supply.
Would you suggest other instrument I could use as a reference?
Offline bktemp
« Reply #4 on: December 04, 2016, 07:06:20 pm »
Unfortunately most available service manuals only cover the older linear regulated power supplies.
If you are lucky the power supply board in HM305-2 is similar.
  • Country: br
« Reply #5 on: December 05, 2016, 04:52:57 pm »
This a photo of the power supply.
Online PA0PBZ
« Reply #6 on: December 05, 2016, 06:33:08 pm »
that looks like a linear supply. Did you look at the link I posted earlier in this thread?
Offline bktemp
« Reply #7 on: December 05, 2016, 06:56:16 pm »
Yes, that is a linear power supply and it looks identical to the one used in HM5010.