Author Topic: Hickok 1605M voltmeter  (Read 979 times)
Offline TimFoxTopic starter
« on: July 14, 2019, 09:40:04 pm »
While I wait for the manual that I ordered on eBay, can anyone tell me the function of each of the six calibration pots facing up on the top board?  I have the version using an AD 545 op amp, not the older version with photo chopper.
Offline coromonadalix
Re: Hickok 1605M voltmeter
« Reply #1 on: July 15, 2019, 01:19:39 am »
I would say take their values  before adjusting them,   and maybe by trial and errors you could find their use ??
« Reply #2 on: July 23, 2019, 08:59:52 pm »
I now have the manual.  In case someone finds this on search, here is the layout of the calibration pots on the top board, seen from above.  Each sets the full-scale value.  Hickok recommends adjusting at 80% of F.S., and using 400 Hz for AC voltage calibration.
         REAR
                    R48 [50 mA DC]
R47
[0.05 V DC]
                    R37
                    [0.5 V AC]
R39
[1.5 V AC]
                    R41
                    [5 V AC]
R44
[50 V AC]
            FRONT