Author Topic: Fluke 865 Graphical Digital Multimeter > needs Calibration  (Read 675 times)

0 Members and 2 Guests are viewing this topic.

Offline sampler1Topic starter

  • Regular Contributor
  • *
  • Posts: 80
  • Country: ca
Fluke 865 Graphical Digital Multimeter > needs Calibration
« on: December 12, 2023, 02:05:59 am »
Hi All,
I have never had any of my gear calibrated so I am comming at you as an ignorant son of a B. haha
I have a Fluke 865 Graphical Digital Multimeter that works perfectly but the DC Volts has slowly gone out of calibration.( roughly +- 1/2 volt)  Does anyone have an idea where I could get it calibrated in Canada? Preferably Quebec? Also what would your guess be on the cost?
I looked in the service manual and I just don't have the calibration equipment to do the job. It does look quite simple.

Cheers and Happy Holidays. :popcorn:
“Electronic devices are powered by white smoke. When smoke goes out, device is dead.”
 

Offline boggis the cat

  • Regular Contributor
  • *
  • Posts: 219
  • Country: nz
Re: Fluke 865 Graphical Digital Multimeter > needs Calibration
« Reply #1 on: December 12, 2023, 02:28:16 am »
Check here for accredited labs:

https://nrc.canada.ca/en/certifications-evaluations-standards/calibration-laboratory-assessment-service/directory-accredited-calibration-laboratories?f%5B0%5D=field_of_calibration%3A30075

There are a fair number in Quebec.  Note that commercial work isn't usually cheap, but possibly if they are too pricey they can suggest alternatives.
 

Offline bdunham7

  • Super Contributor
  • ***
  • Posts: 8010
  • Country: us
Re: Fluke 865 Graphical Digital Multimeter > needs Calibration
« Reply #2 on: December 12, 2023, 04:41:04 am »
0.5 volt on which scale(s)?  Unless that is on the 300V or 1000V range, that is a significant error and I'd be dubious that calibration alone would be the fix.
A 3.5 digit 4.5 digit 5 digit 5.5 digit 6.5 digit 7.5 digit DMM is good enough for most people.
 

Online Swake

  • Frequent Contributor
  • **
  • Posts: 582
  • Country: be
Re: Fluke 865 Graphical Digital Multimeter > needs Calibration
« Reply #3 on: December 12, 2023, 06:41:11 am »
I second that message. If it is half a volt in any lower range it is not a calibration issue unless this has been tampered with. What ranges exactly have that issue?
When it fits stop using the hammer
 