One small hint that can save you a lot of time, if the screen looks like
check if the X and Y plate connectors did fall of and are shorting to ground. It's not easy to get an good picture of the connectors, i hope you can see it in the attached picture.
Offline tautech
Re: Tektronix 2230 repair hint
« Reply #1 on: December 09, 2015, 08:23:11 pm »
Have you checked all the LV PSU rails are to spec for voltage AND ripple?
Unusual trace behaviour like yours PSU is where I'd check first.
This may be of assistance:
http://www.sphere.bc.ca/test/tek-parts/troubleshooting-scopes.pdf
Offline MarkL
« Reply #2 on: December 09, 2015, 08:43:06 pm »
I think the OP is saying the scope is fixed.  The post is a demonstration of what it looks like when a plate connection falls off.
Offline tautech
« Reply #3 on: December 09, 2015, 08:50:21 pm »
Maybe, I'm not sure I read that from the vid.
Anyway hopefully he'll come back and confirm.  :-//
« Reply #4 on: December 10, 2015, 07:49:35 am »
Yes, the scope does work now.