I have recently been working on a Fluke 5200a with a bad case of the power supply flu. Several bad capacitors have caused a cascade of problems through the voltage regulation and main amplifier sections of the device. I'm still in the process of tracking down and repairing all the bad components, but while I have most of the 5200a's guts splayed out on my workbech, I thought some people here might be interested in seeing what makes this thing tick.
<Safety Warning>
Before I begin, its worth mentioning that working on this device requires some added care. While the unit is operational*, the power supply generates lethal potentials of +300 / - 300 Volts. These supply rails are exposed and easily "bumped into" if the case is opened. In summary, the manual puts it best: "DEATH ON CONTACT may result if you do not understand and follow safety procedures."
*In my case, the +190 / -190 V rails were still energized well after power off because the bleed-down resistors were fried.
</Safety Warning>
For those of you who don't already know, the Fluke 5200a is an 53 lbs (24.1 kg) rack-mount AC Voltage Calibrator behemoth. It is able to produce AC voltages in the range between 1mv and 100V at 10Hz to 1MHz. It offers six decades of selectable voltage sub-division and 4 decades of selectable frequency division with a basic AC voltage accuracy of around 500 ppm, which is extendable to the 100 ppm range with an added device-specific characterization and compensation.
Back in the day, the list price was almost 12,000 (USD). Now a days you can pick one up for a few hundred on eBay.
At its core, the Fluke 5200a is a massive servo-controlled amplifier, and by some standards a complete waste of space -- $/gain (main amp) = $4,000 (yup, the main amp's gain is fixed at around 3), power density (watt/lb) = 0.005 (a 741 op amp handles the same power), and completely NON-RoHS compliant (even looking at it for brief time can give you terminal boneitis). But these things (obviously) aren't the reason you'd invest in one of these bad boys. Not when $10 and a smile will buy you one of these guys:
http://www.digikey.com/short/p4p3m.
As many of you with degrees in engineering may find a smile overtly expensive some days, it is worth mentioning that no smile is required to operate a 5200a. All you'll need is an ability to turn knobs -- ALL 19 OF THEM. Which, let's face it, is where the 5200a really shines (only $630/knob!).
Inside the device, you will find 6 or 7 boards that are all connected to a main back-plane. The functionality of the 5200a is logically subdivided onto these boards. Here's the basic theory of operation:
The big goal here is to create a voltage servo-controlled amplifier. The servo loop starts on the AC Oscillator Assembly (purple tabs) where an AC sinusoid is created. This sinusoid is fed into the Power Amplifier Assembly (yellow tabs), where, as previously mentioned, its voltage is increased by a factor of around three. The output of which is fed, by way of a small SMB coax to the Attenuator Assembly (orange tabs) where it is met with an array of precision step-down transformers. The transformer in operation is selected by the range knob on the front of the device (mV, V... etc). If you want the full 100V output, the step-down transformers are bypassed and the main amplifier is cranked to a neck-breaking gain of 30 (gasp!). The output of the active transformer is fed to the front of the device where it is outputted for your use -- diabolical or otherwise. The final output is sensed (internally or externally) and fed into a precision AC->DC converter (see:
https://www.eevblog.com/forum/projects/high-speed-signal-rectification-for-accurate-amplitude-measurement/msg352457/ ). This AC DC Converter Assembly (green tabs) performs a high-speed conversion of the peak AC signal to a DC level. This DC level is compared to another DC level provided by the Reference Assembly (white tabs). This reference assembly is built around the device's DC voltage standard a ye olde SZA263 -- which is really just fancy talk for a temperature controlled buried zener. The error of this comparison, is fed into the Oscillator Control Assembly (blue tabs), which closes the servo loop by directing the AC Oscillator Assembly to correct its output voltage to the main amplifier. VoilĂ ! Couldn't be easier, right?
For the careful observer, the board that has been omitted from the discussion thus far is the RCU Assembly (grey tabs). It is the optional seventh board that allows the 5200a to be remote controlled.
Detailed pictures of each section with commentary will be added and linked to later. Stay tuned.