Hi,
I'm trying to fix up an HP745A AC calibrator. The main square wave generator is an astable multivibrator that feeds a 1kHz square wave in to a bistable multivibrator to produce 30v p-p 500Hz square wave. See first pic. The astable section produces a nice 1Khz 29v p-p square wave but the bistable section generates nothing at all.
It is a frustrating problem. I replaced the three transistors (Q3-5) as they were only 2N3904's and also the diodes. Initially I found that the -25V rail was not working properly but that is fixed now and checked as working. The +/-15v rail, the -25v and +22v rails all seem to be working fine - although the +22v is a little high at nearly +26v.
If I put a scope both sides of CR2 I see the incoming +/- pulses from C4 on one side and only negative rectified pulses on the other side (see 2nd pic) - which is what Id expect. But if I do the same to CR1 I see the incoming +/- pulses on one side but nothing at all on the other side (see 3rd pic). Both diodes are new so I know they are good.
I'm struggling to understand how these rectified pulses disappear for CR1.
Any advice/help would be very welcome!
many thanks
Tony