Hi,
I have a Fluke 189 which was in storage for a long time with batteries in, when I took it out the battery was corroded, I cleaned the battery cartridge, placed new batteries and turned it on the meter was running for itself without measuring anything, probs were in the air and no matter where the selector switch was. I opened the meter to inspect inside found nothing wrong, no battery liquid leaked and it was supper clean, I changed the supper cap, nothing changed in the result, I sent it for calibration they reported “the meter can not register calibration constants.
I tested the eprom 93c56 that holds the calibration constants, it is fine,I can read from and write to it with no problem with an external eprom programmer.
Since it came back from calibration facility, the UNCAL mark is on the screen All the time, and the mere shows 0000 in all conditions, even when I activate calibration mode the measurement is zero, the measurement does not follow the input to match the required voltage calibration menu,
Any one has any thought in that?
Thanks
BM