Ok, it's not the option card.
Manual says this :
CAL: GAIN FAIL (H)
During the CAL AMPTD routine, the amplitude of the video signal at the
ADC on the A16 processor/video assembly is too low. The error message
appears when the reference-level-calibration DAC, located on the A7
analog interface assembly, has been adjusted to its maximum of 255,
and the amplitude of the video signal remains below tolerance.
Related Assemblies
8590L and 8591E spectrum analyzers only: A3 input attenuator, A4
first converter, A5 second converter, A6 YTO, A7 analog interface, A9
third converter, A11 bandwidth filter, A12 amplitude control, A13
bandwidth filter, A14 log amplifier.
All other 8590 E-Series and L-Series spectrum analyzers: A3 front end,
A7 analog interface, A9 third converter, A11 bandwidth filter, A12
amplitude control, A13 bandwidth filter, A14 log amplifier.
Troubleshooting Hints
The CAL AMPTD routine checks the amplitude of the video signal as it
adjusts the A7 reference-level-calibration DAC. Normally, the
video-signal amplitude should move within tolerance as the A7 DAC
changes the gain of the IF calibration amplifier on the A9 third
converter assembly. Figure 9-1, IF/Control Overall Block Diagram, and
Figure 9-2, Figure 9-3, or Figure 9-4, RF Section Overall Block Diagram
for more information.
Check the signal amplitude along the signal path from the RF input,
through the IF section, to the output of the A14 log amplifier. Refer to
“Isolating an RF, LO, IF, or Video Problem” in Chapter 4 for further
troubleshooting information.