Author Topic: EZ OS-5020 Oscilloscope Not working right.  (Read 3824 times)

0 Members and 1 Guest are viewing this topic.

Offline coolduder1001Topic starter

  • Contributor
  • Posts: 10
  • Country: us
EZ OS-5020 Oscilloscope Not working right.
« on: August 08, 2021, 09:21:25 pm »
This is my first oscilloscope and I've already run into an issue.  When I tried to use it for the first time I don't get a line across the screen, and when I wired CH1 to to the test point next to the power switch I could just faintly see something moving in the line.  It's almost like everything in the vertical space was squished into the middle.  Also the intensity knob and X pos. knob do nothing.  I took the scope apart to see if anything was obviously blown and I can't find any issues, other than one resistor (r428) has a little discoloration at it's joints.  Other than that I don't see any obvious issues and I don't really know where to go from here.

I have a folder here with pictures of the MB and a video of me trying to operate the front: https://drive.google.com/drive/folders/1JYNL4ASaM53rDvzh0avbxfCi7vpnpRsL?usp=sharing

Thank you for any help.
« Last Edit: August 09, 2021, 01:07:43 am by coolduder1001 »
 

Offline wn1fju

  • Frequent Contributor
  • **
  • Posts: 572
  • Country: us
Re: EZ OS-5020 Oscilloscope Not working right.
« Reply #1 on: August 10, 2021, 12:28:01 am »
1.  Try to find a service manual online with schematics.
2.  A couple of your pictures seem to show a connector that isn't plugged in (did you disconnect it intentionally for the photo?).
3.  The first step in the repair of pretty much any piece of test equipment is to check the power supply voltages.  Look for sagging DC voltages and/or excessive ripple on each supply rail.  There is a good chance that one or more electrolytic capacitors have failed.
4.  Be careful where you stick your hands.  The CRT uses enough high voltage to zap you pretty good.  If you don't know what you are doing, STOP.
5.  If all the supply voltages check out OK, you have to start debugging the various vertical and/or horizontal circuits.  Unfortunately, the best way to do this is with a separate (working) oscilloscope.

 

Offline coolduder1001Topic starter

  • Contributor
  • Posts: 10
  • Country: us
Re: EZ OS-5020 Oscilloscope Not working right.
« Reply #2 on: August 10, 2021, 04:40:30 am »
1. I tried finding schematics for this scope with no luck.  I found one for the os-9020 which is completely different since the horizontal and vertical have dedicated boards on that model.  All I could find was the manual for the scope https://www.testequipmentdepot.com/ezdigital/pdf/os5020-5020c_manu.pdf

2. Yes, I forget to put that back.  I was checking the power supply voltages while it was disconnected, I forgot to write down the readings.

3. I did check the power supply voltages while all the wires were connected and I have it starting at R801A going down.  From what I could find these are not the correct voltages but I can't find out what is.  Someone has a similar issue but with different voltages https://www.electronicspoint.com/forums/threads/repair-help-on-lg-os-5020c-oscilloscope.276494/ .  I did check the capacitors on the power supply board with my component tester and they were all within spec with no bulging or leaking.  From what I can remember running the board disconnected from the mainboard the voltages looked more normal, but I still have no ide if they were.

1 - (260v)
2 - (5v) -This one should be correct
3 - (20v)
4 - (GND) -This one should be correct
5 - (-20v)
6 - (0v)
7 - (210v)
8 - (0v)

4. I know CRTs are scary so I'm avoiding it as much as possible. I don't think it's an issue with the tube itself anyway.

5. Well the power isn't ok sooooo...

PS: I am a bit concerned about that resistor (R428) next to the connector I disconnected.  It looks like it got really hot and discolored the board a little.  I want to look for replacement just in case, but I can't really tell what kind it is.  I also can't tell the color to good so I can't read the color code.
« Last Edit: August 10, 2021, 04:45:09 am by coolduder1001 »
 

Offline wn1fju

  • Frequent Contributor
  • **
  • Posts: 572
  • Country: us
Re: EZ OS-5020 Oscilloscope Not working right.
« Reply #3 on: August 10, 2021, 11:47:15 am »
It is probably very similar to the OS-9020, so I would use those voltages as the "normal." 

All of your voltages look quite a bit high (except the ones that are at zero!).  For instance, you should be getting +12 and -12, not +20 and -20 volts.

I would disconnect all the P/S connectors to the boards and then check the voltages being produced.  Check them for DC and for AC.  Excessive AC points to an electrolytic and/or diode problem.  Next, I would check the external (I think there are 3 of them) pass transistors for shorts.  For instance if you have 20V being input to the 12V regulator circuit via the pass transistor, and the transistor is shorted, you could get 20V out.  Note:  I had a similar EZ-scope and was getting 15 and -17.5V instead of 12 and -12V.  All I got was a small circular trace on the left edge of the screen.  Once I corrected those voltages (pass transistor short), the scope worked fine.

If all the open circuit voltages are OK, then go the corresponding connectors on the various boards and measure the resistance to ground for each pin.  You want to know if there is something on a board that is dragging down the P/S.

It may be that the discolored resistor is a low ohm current sense resistor.  If the voltages are too high, or the current draw is too much, and the regulators aren't working, you could stress that part.  Most overheated resistors I've seen actually measure fine, so measure it with an ohmmeter and see what you get.

I wouldn't run the scope for too long with those measured voltages.  They may inadvertently stress some other parts.

 

Offline Chris56000

  • Frequent Contributor
  • **
  • Posts: 909
  • Country: gb
Re: EZ OS-5020 Oscilloscope Not working right.
« Reply #4 on: August 10, 2021, 07:02:12 pm »
Hi!

I posted the schematics for this Oscilloscope about two or three weeks ago!

https://www.eevblog.com/forum/repair/htc-os-5030-oscilloscope/

Chris Williams
It's an enigma that's what it is!! This thing's not fixed because it doesn't want to be fixed!!
 

Offline coolduder1001Topic starter

  • Contributor
  • Posts: 10
  • Country: us
Re: EZ OS-5020 Oscilloscope Not working right.
« Reply #5 on: August 11, 2021, 03:14:25 am »
Ok I checked the voltages again with the power supply disconnected.
1 - (230v)
2 - (5v)
3 - (22v)
4 - (GND)
5 - (-22v)
6 - (0v) -----
7 - (150v)   |--- (2.5v AC)
8 - (0v) -----

The 7805 (15v in 5v out), 7812 (20v in 12v out), 7912 (-20v in -12v out).  The last chip on the heatsink is a 2SA1668 transistor and I don't know much about how it's wired, I don't see any voltages on it when it's on
 

Offline perieanuo

  • Frequent Contributor
  • **
  • Posts: 914
  • Country: fr
Re: EZ OS-5020 Oscilloscope Not working right.
« Reply #6 on: August 11, 2021, 05:17:39 am »
hi, from op seems your vertical deflection is dead, trace the power supply for vertical board
so focus first on this vertical board i see in sm for OS-9020G page 60 the voltages are +-12VDC, so maybe this module or his input are the culprit (i didn't watched the video but the photo says V deflection is dead)
higher voltages means somebody doesn't operate and of course V goes higher
i'm not sure the schematics are correct for your model, but may be used as guidance for your case, usually they just 'tune' a little bit between models of same product series.
please refer to that manual if you can, it's simpler for us to trace this than to trace R801A or R428 in some old pdf, we can't search text in that pdf, at least not after 1 hour of ocr work on that pdf.
or draw on paper what you reverse engineer
[edit]
now i'm sure the schematic 134639836-LG-OS-5020G-Oscilloscope-Service-Manual.pdf is not the same, you have some picture with P210 connector and R428 comming from pin 6/P210 and in the schematic page 59 we have R428 2K tied to +12V rail. if any resistor is fried there should be R429, but in that V defl preamp i doubt resistors are hot.
i'll dig a little more see if i got something valuable :)
« Last Edit: August 11, 2021, 05:30:00 am by perieanuo »
 

Offline coolduder1001Topic starter

  • Contributor
  • Posts: 10
  • Country: us
Re: EZ OS-5020 Oscilloscope Not working right.
« Reply #7 on: August 11, 2021, 05:22:30 am »
I just compared it to the os-9020 diagrams and they're the same.  I can tell cause the power board for both the one you sent (465F8-8.PDF) and the lg os-9020 are a perfect match.  Plus on my board there is only one 7805 on the PS itself and no transistors.  Both of these are also missing part (NL801)  I have no idea what this component is.  It looks kind of like a Christmas light with two metal rods in it, it also flashes when you turn the scope on and I don't know if it's supposed to do that.
 

Offline coolduder1001Topic starter

  • Contributor
  • Posts: 10
  • Country: us
Re: EZ OS-5020 Oscilloscope Not working right.
« Reply #8 on: August 11, 2021, 05:25:53 am »
I don't think this os-9020 schematic is helpful at all.  Because my PS does output +-20v, but the voltage regulators are moved to the main board except for the 7805.  The 7812, 7912, and 2s1668, are all on the mainboard.

I just added some close-ups of the PS board to the folder.
« Last Edit: August 11, 2021, 06:17:46 am by coolduder1001 »
 

Offline perieanuo

  • Frequent Contributor
  • **
  • Posts: 914
  • Country: fr
Re: EZ OS-5020 Oscilloscope Not working right.
« Reply #9 on: August 11, 2021, 05:50:33 am »
if you think schematic is ok, do you have the right voltages see page 67 Power Supply (14)?
vertical deflection preamp/amp works with those +-12Vdc,+140Vdc, +5Vdc see connector W705
and btw, +-12V are regulated voltages, founding +-20V instead is not ok at all
 

Offline Labrat101

  • Regular
  • **
  • Posts: 688
  • Country: 00
  • Renovating Old Test Equipment & Calibration ..
Re: EZ OS-5020 Oscilloscope Not working right.
« Reply #10 on: August 11, 2021, 05:54:39 am »
 It looks kind of like a Christmas light .
It's a spark gap . Or neon stabilizer . Used to stop high voltage surges.  Momentary flash at power on is not uncommon.
"   All Started With A BIG Bang!! .  .   & Magic Smoke  ".
 
The following users thanked this post: coolduder1001

Offline coolduder1001Topic starter

  • Contributor
  • Posts: 10
  • Country: us
Re: EZ OS-5020 Oscilloscope Not working right.
« Reply #11 on: August 11, 2021, 06:19:46 am »
I said it's not helpful.  Look at the pictures of the PS board and tell me if that matches the schematic.
 

Offline Labrat101

  • Regular
  • **
  • Posts: 688
  • Country: 00
  • Renovating Old Test Equipment & Calibration ..
Re: EZ OS-5020 Oscilloscope Not working right.
« Reply #12 on: August 11, 2021, 06:46:58 am »
As mentioned above .
You have a bad ceramic cap in the vertical . . The circuit diagrams.  Is not complete .
I will look tonight when I have a large screen to look closer.
Also what voltage are you . 115v or 240v ?
"   All Started With A BIG Bang!! .  .   & Magic Smoke  ".
 

Offline perieanuo

  • Frequent Contributor
  • **
  • Posts: 914
  • Country: fr
Re: EZ OS-5020 Oscilloscope Not working right.
« Reply #13 on: August 11, 2021, 07:07:37 am »
I said it's not helpful.  Look at the pictures of the PS board and tell me if that matches the schematic.
it is helpful, it doesn't need to be identical, is totally ok for guidance
i always pointed you
Quote
you have some picture with P210 connector and R428 coming from pin 6/P210 and in the schematic page 59 we have R428 2K tied to +12V rail. if any resistor is fried there should be R429, but in that V defl preamp i doubt resistors are hot.
, your board and 9020 schematic are a little different, but that doesn't mean you can't use those schematics to find the problem
someone with 5020 found different voltages https://www.edaboard.com/threads/repair-help-on-lg-os-5020c-oscilloscope.347582/ , CHECK PS like i said on my last post and report
don't search for 'bad ceramic cap' or any other gizmo before validating 100% the supply voltages, and so far you haven't found the correct values expected. each thing at the time, learn the correct approach to repair electronics
 

Offline coolduder1001Topic starter

  • Contributor
  • Posts: 10
  • Country: us
Re: EZ OS-5020 Oscilloscope Not working right.
« Reply #14 on: August 11, 2021, 07:51:12 am »
I'm using 115v.
 

Offline coolduder1001Topic starter

  • Contributor
  • Posts: 10
  • Country: us
Re: EZ OS-5020 Oscilloscope Not working right.
« Reply #15 on: August 11, 2021, 08:01:52 am »

CHECK PS like i said on my last post and report
[/quote]

That's what this is.
1 - (230v)
2 - (5v)
3 - (22v)
4 - (GND)
5 - (-22v)
6 - (0v) -----
7 - (150v)   |--- (2.5v AC)
8 - (0v) -----

These are the voltages I got when I disconnected the Power Supply board from everything else in the case.  I can't see how the components on this board will give me a +-12v.  Unless the transformer after it's rectified gives out a perfect +-12v. 
 

Offline perieanuo

  • Frequent Contributor
  • **
  • Posts: 914
  • Country: fr
Re: EZ OS-5020 Oscilloscope Not working right.
« Reply #16 on: August 11, 2021, 08:12:37 am »
it doesn't matter 220 or 110 or 50/60hz
i repeat, page 67, you have to check those voltages or at least try to reverse that part, for example diode D924 will tell you the correct regulated voltage, i doubt they switched from 12 to 20V, but anything is possible
 

Offline perieanuo

  • Frequent Contributor
  • **
  • Posts: 914
  • Country: fr
Re: EZ OS-5020 Oscilloscope Not working right.
« Reply #17 on: August 11, 2021, 08:15:03 am »
btw, i don't see that +55Vdc they mention in page 67, just near the +140Vdc
 

Offline ambrosia heart

  • Regular Contributor
  • *
  • !
  • Posts: 133
  • Country: hk
Re: EZ OS-5020 Oscilloscope Not working right.
« Reply #18 on: August 11, 2021, 08:37:15 am »
LG = EZ OS 5020.
Comparing voltages and see photos.
 

Offline Labrat101

  • Regular
  • **
  • Posts: 688
  • Country: 00
  • Renovating Old Test Equipment & Calibration ..
Re: EZ OS-5020 Oscilloscope Not working right.
« Reply #19 on: August 11, 2021, 11:48:06 am »
This is the service Manual for your Scope . This scope was build by LG and was rebranded
Also known:
LG PRECISION GOLDSTAR OS-5020 OS5020 OS 5020
https://elektrotanya.com/lg_precision_goldstar_os-5020_oscilloscope_sm.pdf/download.html#dl

Please enjoy . All your Power supply voltages are out . and the 56 volt could well be your problem .

PS. I am not sure what country you are from . But please check your Grammar before posting as you have given quite a few negative answers . We are all he to help . We can't actual see what's on your bench and what test equipment you have at hand . I am also assuming you are a novice in electronic by some of your questions . Yes it is frustrating when something  goes wrong .
As already said the power supply is the first priority .
There are also differences in Models as in version numbers.  So some components maybe  changed due versions update or batch

https://elektrotanya.com/lg_precision_goldstar_os-5020_oscilloscope_sm.pdf/download.html#dl
« Last Edit: August 11, 2021, 12:18:17 pm by Labrat101 »
"   All Started With A BIG Bang!! .  .   & Magic Smoke  ".
 

Offline ambrosia heart

  • Regular Contributor
  • *
  • !
  • Posts: 133
  • Country: hk
Re: EZ OS-5020 Oscilloscope Not working right.
« Reply #20 on: August 11, 2021, 12:12:08 pm »
 

Offline Labrat101

  • Regular
  • **
  • Posts: 688
  • Country: 00
  • Renovating Old Test Equipment & Calibration ..
Re: EZ OS-5020 Oscilloscope Not working right.
« Reply #21 on: August 11, 2021, 12:21:32 pm »
"   All Started With A BIG Bang!! .  .   & Magic Smoke  ".
 

Offline coolduder1001Topic starter

  • Contributor
  • Posts: 10
  • Country: us
Re: EZ OS-5020 Oscilloscope Not working right.
« Reply #22 on: August 11, 2021, 05:49:37 pm »
WOW it looks like you have the actual manual, I can't find that thing anywhere.  I just get people sending me the os-9020 manual thinking it's the os-5020.
 

Offline Labrat101

  • Regular
  • **
  • Posts: 688
  • Country: 00
  • Renovating Old Test Equipment & Calibration ..
Re: EZ OS-5020 Oscilloscope Not working right.
« Reply #23 on: August 11, 2021, 06:59:18 pm »
WOW it looks like you have the actual manual, I can't find that thing anywhere.  I just get people sending me the os-9020 manual thinking it's the os-5020.
9020 read page 3  your scope is upgradable the PCB is similar ie as OS-9000 ,3000srs series
If you read page 3 it has the versions and corrections list . and models covered
Please Note the schematic is correct but you will find component numbers may differ . due to version updates .
Also  the AC board is not shown as being separated in the schematic . The main power control is on the main board .
Sorry you will have to do a bit of trace work . 
There will also be components layout differences  ie your Hot resistor 428 on some models its 2 resistors in parallel  .
I hope this helps . also read the section explaining how sections work .

Have fun & good luck

PS  update I would check the 2SA1668   as this supplies the Vertical voltages . This transistor has probably gone leaky  or failed causing the resistor
R428 to run Hot
« Last Edit: August 11, 2021, 07:40:10 pm by Labrat101 »
"   All Started With A BIG Bang!! .  .   & Magic Smoke  ".
 

Offline coolduder1001Topic starter

  • Contributor
  • Posts: 10
  • Country: us
Re: EZ OS-5020 Oscilloscope Not working right.
« Reply #24 on: August 11, 2021, 10:12:17 pm »
That's kinda what I guessed that's why I mentioned it here.

The last chip on the heatsink is a 2SA1668 transistor and I don't know much about how it's wired, I don't see any voltages on it when it's on
 


Share me