Welcome to the forums
I would say it's not the fusible surge resistor R1 1kΩ or the PTC causing the open-circuit Ohms reading 10.5MEG because they are so small in comparison to the input resistance and their purpose is to protect the Ohms-source. So if they vary 20% it's still nothing for a couple kΩ to the 10MEG in the divider.
Usually it's a damaged protection clamp transistor Q1 or Q2 that is leaky. You can lift one leg to disconnect them and see if the reading persists.
It rarely can be the protection MOV's leaking RV2, RV1.
A dirty PCB or rotary switch, especially around the Z1 9.999MEG, Q1&Q2, across RV1,2 etc. Use IPA and a Q-tip to clean up old flux and any battery leakage etc. and let it thoroughly dry out.
Beyond this I think the DMM IC can get zapped and a damaged CMOS MUX internal switch give the leaky Ohms reading.
Another way to flush out the leaky part is to apply heat - with your finger, soldering iron etc. and find what makes the reading jump significantly.