I measured the input protection and it seems fine. RT1 measured 1.31k; R20 was 1.05k so in spec and the three RV's were all OL as you said.
All the above readings suggest nothing wrong.
I think there is some leakage somewhere.
I have to agree. The current reading for 400 ohm range seems really low to me. Of course, the manual doesn't specify, but I'm guessing that is low?
Shorted probes in VDC gives a reading of 0.004V and millivolt mode gives 003.9mV.
Is the input impedance still 11.11M ohm for VDC and NOT 10.00M ohm for mVDC?
The more I clean it the worse it seems to get!
Try drying the pcb more with a pc fan and see if the ohms readings gets better.
For some reason, my brain is nagging me that I have seen this before somewhere, but I can't recall what fixed it or where the thread might be in eevblog.
I think it was a Fluke meter, but I can't recall the circumstances on how it was fixed regarding the ohms offset.
Let me sleep on it and see if I can remember in the next couple of days?
Another question/test. Does diode work properly?