Bonjour Shakalnokturn,
Please note for the clarity that I've two TDS540C in my lab, one which I've purchased on eBay working fine and another one which is partially failed.
This morning I've removed as you suggest the RS-232 and Centronics optional board from the main A11 board. When starting the scope, now it will not display the Option 13 on the banner so it seems the TDS540C self-detect by hardware the presence of Option 13. Unfortunately I still have FAIL - Processor so there must really something on the A11 processor/display board.
By the way, I notice a strange thing when running side by side the healthy TDS540C (2GS/s) versus the failed TDS540C (1GS/s option 1G). When I push to the limits the horizontal scales up to 50ns, the left-top screen will correctly display " tek Run 1GS/s " but if increasing the horizontal speed, the healthy TDS540C will always keep tek run 1GS/s so this strange because normally it should offer 2GS/s standard. However on the partial fail TDS540C Option 1G, I can push to faster horizontal speed up to 100 GS/s at 50 ps except it displays additional ET SAMPLE
So one TDS540C has printed 500MHz - 2GS/s whereas the other 500 MHz - 1GS/s Option 1G.... which one is fastest after all ?
Anyway sorry for the rambling where the topic is understand how to eventually repair the A11 process or fail which by the ways freezes the SPC initialized.
Lastly I really confirm the size of the log file is not the problem, each time I re-run the scope it will add same error estimates with new date and time but keep the oldest log file data from 2003. I'm attaching the historical logfile from 2003 where it seems this Oscope had many problems in the past then kept sitting alone with the last problem, namely Processor failed.