A few pictures and some measurements of the front end on a 500MHz unit.
Awersome.
Thank you very much!
The inductances are too low for the meter to measure.
Yes, but at 100 kHz these inductors can be measured without problems.
I also made some measurements in the front end of an MSO6104, I attached the results, below.
I used a DER EE DE-5000, calibrated according to the different sizes (603, 1206), to compensate the deviation introduced by the SMD test clips.
The parts marked in yellow can be measured on the board, without remove them. The other components must be removed from the board to get its real value.
Note 1: The first network performs filtering and attenuation to the ranges below than 500mV (IN_1, OUT_1). It's awesome, Agilent did a good job with these machines.
Note 2: According to the inductor's datasheet it's DC resistance increases proportionally to its inductance, but this is not met by the 100nH inductor, I don't know why.
For 33nH -> DCR = 1.35 OHM.
For 47nH -> DCR = 1.71 OHM.
For 100nH -> DCR = 0.7 OHM.
In a few days I will try to do the same with my MSO6034A.
Cheers!