Author Topic: [SOLVED] Agilent 54502A CRT display broken  (Read 1917 times)
Offline brych34Topic starter
  • Country: wales
« on: September 26, 2019, 05:05:58 pm »
Hello everyone
I'm a newcomer to electronics and by luck I recently acquired an old 54502A 400MHz digitizing oscilloscope that was being thrown away: https://ibb.co/M9bfnMp
Unfortunately the cathode ray tube display is broken. When I use the 'self-test' function the entire display is supposed to light up, yet only part of it down the centre does: https://ibb.co/hKXV2F2
If it's just a case of replacing a component I'll try and repair it but if you think the display itself is busted I'll take it apart for parts.
Thanks for the help
« Last Edit: February 23, 2020, 09:14:13 pm by brych34 »
Offline SoundTech-LG
Re: Agilent 54502A CRT display broken
« Reply #1 on: September 27, 2019, 01:38:36 pm »
That CRT display looks great. Your problem is likely in the horizontal deflection. Likely the vertical mounted board to the left of the CRT. Unfortunately I don't think there is a schematic for it.
The following users thanked this post: brych34
  • Country: wales
« Reply #2 on: September 27, 2019, 04:16:21 pm »
Thanks - I'll have a look at it tonight after I discharge the display.
Offline TurboTom
« Reply #3 on: September 27, 2019, 08:00:09 pm »
These monochrome CRT displays usually run their horizontal deflection through a quite substantial bipolar electrolytic capacitor of a few microfarads. Over the decades, these caps will dry and ESR will increase until the horizontal deflection fails (almost) completely. You may want to search for such a capacitor on the CRT PCB and check its capacitance / ESR. Other components in its vicinity may fail as a consequence.
The following users thanked this post: fcb, brych34
  • Country: wales
« Reply #4 on: September 28, 2019, 12:02:00 pm »
I'll see if I can get my hands on an LCR meter. Thanks
  • Country: wales
« Reply #5 on: February 23, 2020, 09:11:38 pm »
It works now! I replaced the 3.3uF bipolar capacitor on the board to the left of the CRT with one from AliExpress:
As far as I can tell it works perfectly now:
Thank you!
Offline GreyWoolfe
  • NW0LF
Re: [SOLVED] Agilent 54502A CRT display broken
« Reply #6 on: February 24, 2020, 01:39:49 pm »
Offline fcb
    • Electron Plus
« Reply #7 on: February 24, 2020, 02:19:38 pm »
Excellent repair. These are great scopes - especially for free.
Remember they are equivalent time sampling at 400MHz. Realtime to 100MHz.  But still excellent!
« Reply #8 on: February 24, 2020, 09:21:09 pm »
Great Job!  I'll try to remember that fix. Lots of 54xxx scopes, logic analyzers, and the 53310A Modulation Domain Analyzer use that same CRT/board combo. How did you happen to find the cap???
Offline TK
« Reply #9 on: February 24, 2020, 09:38:53 pm »
I had the same issue with an HP8714B and replaced the bipolar capacitor.  They come in a much smaller footprint now and it worked when I replaced it... but then later another magic smoke and the new capacitor was gone.  I asked the forum and I remember somebody recommended using a non electrolytic capacitor and that is what I did and it has been working for 3-4 years now.
  • Country: wales
« Reply #10 on: April 09, 2020, 04:14:05 pm »
Sorry for the late reply. I just searched for "3.3uF bipolar capacitor" on AliExpress and it led me to here: https://www.aliexpress.com/item/32594169451.html
So far it seems to be working well!
The following users thanked this post: mightyohm