Used a 9V PSU
Tried every range on VDC and values don't change.
re measured μΑ on ranges after cleaning the flux and doing the range change.
Results :
000. ohm range -999.1 μA
0.1k range - 999.3μA
00.1k range -99.46μA
000.1k range -9.84μA
0.1M range -4.83μΑ
00,1M range - 0.36μΑ **
000,1Μ range -0,36μΑ **
The reference currents are all still n.OK, i.e. about 140..180nA too low, which is clearly visible for the 500nA, 5µA, 10µA and probably also for the 100µA reference currents.That's the reason, why you get ERR 612, 618, which both indicate a failure of the 500nA current, i.e. out of tolerance.
U201 was really defective, as initially none of the expected voltages @ pin5, 6 were ok, instead showing the usual faulty values for too high bias currents, i.e. broken input stages.
Now you have to search, where these about 180nA are flowing to GND.
Hopefully your new U201 is OK, then Q211 would be the next suspect. So let's at first rule out both these failure modes:
Please desolder CR202, and then measure the reference current directly in situ towards AGND, i.e. from pin 3 / collector of Q209 to AGND, which is @ INPUT LO.
you might check for the lower currents of 500nA, 5µA and 10µA only, so you have to select 10M, 1M and 100k ranges (see page 98 of service manual). An external resistor is not needed.
Please compare these three currents with those you have measured in the above table.
Frank