Author Topic: 34401a 601 error repair  (Read 739 times)
Offline f14Topic starter
  • Country: vn
« on: May 18, 2022, 01:20:54 pm »
Hi Guys
I have agilent 34401a with frontpanel damage. VFD display is broken and I buy a new one(clone ) from aliexpess.  after install new VFD . my 34401a have two beep ( the first one is quiter a bit) and VFD display is bright all segments and suck in this. all buttons is not respond
I remove the display board and connect the mainboard to PC via GPIB and this mainboard working correctly . (just have one error 601) all functions  are good
so I think display board  had damage too . I have check LM339M but it normal. seem the MCU 80C51 is dead ?
thank you!
« Last Edit: May 18, 2022, 01:25:31 pm by f14 »
Offline strawberry
  • Country: lv
Re: 34401a 601 error repair
« Reply #1 on: May 18, 2022, 04:19:15 pm »
broken trace or solder
  • Country: vn
« Reply #2 on: May 19, 2022, 12:14:32 pm »
I realize N80C51BH is dead. so I replace it by AT89S52  with fw found in here . after that display have this message "QTTR022" and when I perform self test it show like second picture
when I connect the display board with mainboard. I can not access controll via gpib . it freeze . have any idea for this issue ?
solved. this fw 88813 is too "new" for my 34401. 88804 working