You've inspired me to try my hand at this, since I've had all the parts for it in the drawer already.
I've used GDR-made SMY60 (dual P-FET without gate protection zeners) as a input stage into an OP07, 1G KVM in the feedback.
The 3pF feedback capacitor is made of three 0805 NP0 10pF in series (is that a good idea?).
I've just now made and set it up:
- shorting the input (wondering why the output pegs
![Face Palm :palm:](https://www.eevblog.com/forum/Smileys/default/facepalm.gif)
- measuring at the two inputs of the opamp
- then set the balance potentiometer of the input stage to get zero across -IN +IN.
To try it out:
- remove short, put 100Meg towards the bias voltage source on the outside
- feed in 1V, expect -10V out.
But out come only -7.42V. With -1V in, the output is 7.42V. With no input, output is pretty much zero
![Scared :scared:](https://www.eevblog.com/forum/Smileys/default/scared.gif)
Possible conclusion to me:
- Feedback cap is leaky, shunting the 1G
- 1G is lower or dirty
- 100M is higher (metal film, this much drift up seems less likely?)
Is the feedback cap the likely culprit, or is it dirt on the 1G? Or something else entirely?
Edit: nevermind, just realised it's oscillating, 3pF seems too little.
Edit#2: with 100pF PS foil capacitor, after waiting a loooong time (for the capacitor to fully dry in circuit), it settles at 7.41V. So not the capacitor leakage?
Edit#3: These 100M all read high
![WTF? :wtf:](https://www.eevblog.com/forum/Smileys/default/wtf2.gif)
The one used measures 120.8Meg. With 1.208V Bias, -8.96V comes out. (-1.208V -> 8.97V) Still "a little" short, but closer to reasonable.
Edit#4: At the same time / from the same ebay seller, I got other 1G, an 100M, etc. among other values. After quick rinse with IPA, the 100Meg reads low at about 90M, and the other 1G reads ~905M (even though I have less confidence in the latter measurement). Coincidence, or did these KVM all drift low with age? I should have measured the one I put in beforehand
![Banging Head |O](https://www.eevblog.com/forum/Smileys/default/bangheadonwall.gif)
Edit#5: With 120.8Meg & -12.08mV Bias the output is 94.5mV, at +12.08mV the output is 85mV. Is that explaineable as ~5pA leakage?
Edit#6: With 0.57mV Bias on the 120.8Meg, the output is zero, so ~5pA input leakage confirmed?