If the MSD op-amp has a 1K resistor to the summing amp, the others have 10K, 100K, 1M, 10M and 100M for the LSD.
Liked it! Nice design, but isn't the 100M resistor a bit high? Isn't a little bit better to use a 0.1 hamon divider with it´s output a little higher (about 0,1%) followed by a 10M resistor?
100M is a bit high, and using a 10:1 divider or a 100:1 is a great solution. Accuracy is not important - you can calibrate that out. Stability is important. One thing you can do is to use the divider networks that have great temperature matching, so you get good stability. Since it is a one-off build, it is perfectly reasonable to manually select resistor combinations until it is very close to correct, then use a pot to adjust for the last 1 part in 100 of accuracy. Also the high resistances are for low order decades, so an an accuracy of 1% is adequate. You do not need to go overboard. It is the first 4 decades that you have to be obsessive about.
I think easy calibration is the way to go for the average bear. What I don't know is the magnitude of the error sources that "come out in the wash" when the Analogic unit is trimmed. That trimming might have compensated for switch resistance, leakage and who knows what else. The unit did step in 10 uV steps on the 10V range. IMO, one of the nice things about the way they did it was being able to go a bit over 1 or 10V; It's very common to need that feature. I like the idea of modules for each decade.
Easy calibration is everything I hope for!
On going a bit over 10V, Richard's design can do that too. If you select the 10V and the 1V outputs, you get 11V. With 6 decades, you can go to 11.1111V.
Good ideas with the modularisation of the design. I'll keep it in mind.
I wasn't sure if my explanation for the easy calibration method was clear, but if you followed it, then great!.
I stumbled across these switches that look like they might be OK for a Kelvin-Varley divider. 2 pole 11 positions (which I prefer to 10 position).
Nice find! But why 11 positions are better? Don't you only need 10 positions for a KVD?
Thank you,
Felipe
To calibrate the resistors in a KVD stage, you need to be able to break the connection between at least one leg of the chain and the rest of the KVD. Going all the way with the Fluke 720a system with the whole calibration for the first 3 stages managed by switches is the ultimate, but they can afford to buy great switches - and great switches can easily cost hundreds of dollars. Having an 11th position that breaks the chain free of the next lower divider, combined with a calibrate procedure that perhaps using plugable wire links rather then Fluke's switches, is a workable solution for a home builder. There would be other ways to do it I am sure, and some people will just hate having that 11th position.
There still need to be another switch that can disconnect all the resistor in parallel with the dividers, or maybe not with guarding during calibration? Exactly how how to make it all work will need some thinking.
Calibrating the KVD is all about keeping the box sealed and untouched as much as possible. If there was a removable panel that gave access to the adjustment pots, and the pins that you could link for calibration, that would be ideal.
You can safely match resistors to 0.01% before assembly, so like the Fluke, it is best if the lower 4 ranges have no calibration at all. That means they really only need 10 position switches. Just have to find good ones at a decent price. The supply of affordable good quality rotary switches is not what it used to be.
On the first decade, an 11th position is essential, as it means you can you can have a "1.1" input like the Fluke 720A. This allows you to do plus-minus variations around the 10.00000 mark. But you also need to be able to break the connection to the second divider, so does that mean we need a 12 way? Very possibly.
It is very annoying if you can only go to 9.99999 and no higher.
Richard.