hmmm well this is interesting, it seems both are genuine. they both exhibit the inrush (not as high as speced but approx 60-120mA depending on how cold is the startup). and thanks branadic for the pic, it seems your NS variant capsule looks like the way it is as the used units i have gotten, the plastic flag to denote the pin orientation is not very obvious like LT. the insides of these pieces are also gold, so they have re-tinned almost the entire lead. are the LT variant made in a different process with better material
? their leads look so "solid"
and the plastic looks very "fine" cosmetically ?
@andreas : yea the plastic can are really very rattly loose. when poking them onto the bread board, 1 of them decided to come out and say hello ! then i saw the metal can, no markings, bottom side are gold
when tested @ 27v
unit A with 6.95v settles at about 10.7mA (heater current only). case temp 63
oC
unit B with 6.81v settles at about 12.6mA (heater current only). case temp 67
oC
both are fed 0.98mA into the zener (20k feeder), both drops near to the stable current very quickly in about 3-4s
when the zener feed is increased to 1.9mA, both heater current also increases by approx 1mA, i can see the there is a diode connection between the zener side to heater, i guess the zener feed does affect the heater somehow
? hmmm how could it be ?
i guess i have nothing else for short term tests until i figure out a way to "SCPI" the used keithley thru the RS232, then the routine starts to try and "see" the drifts
. im having such a high hill to climb learning to use SCPI, this is going to take a long time
(maybe LM399 sellers should give a longer timeframe for confirming a purchase? lolz, tough business?!)
again thanks folks for all the tips pics and helps