Studying the effects of bias voltage on capacitors usually requires an LCR meter with a build-in bias network, or configuring an external network to isolate the DC bias from the LCR meter terminals which has no built-in bias or the LCR bias doesn't have sufficient range.
With a brief break in other work we decided to create an external network to support external DC bias with the bench type LCR meters we have and use. After some careful experimenting we found the Tonghui TH2830 to be receptive to AC coupling the High Current and High Potential terminals, our other lab bench meter wasn't tho!!
With a relatively large range of potential test capacitors, which includes ceramic, electrolytic, tantalum, and polymer, and a large voltage range we decide to investigate the HP16065A (service manual kindly provided by Thomas) External Voltage Bias Fixture. This provided a baseline schematic and methodology to move forward.
Here's our first cut at an external bias fixture to use with a standard bench type LCR meter. The general idea is to insert the fixture between the LCR meter terminals and the actual component testing fixture, wether it be a set of Kelvin clips, tweezers, SMD holder or leaded device fixture/holder. With the standard spacing of the BNC terminals the bias fixture can be directly connect to the LCR terminals with 4 BNC male to male adapters, or with quality BNC cables (either 50 or 75 ohm). The component fixtures attaches to the other side of the bias fixture.
In the schematic the LCR High Current and High Potential terminals are isolated with series capacitors. The design is flexible to allow paralleling film type capacitors for greater capacitance, and the Current terminal also supports high capacitance electrolytic types as shown. The LCR terminal are protected with back to back diodes, back to back Zener diodes, and anti-parallel series Zener diodes with a GP diode. Also the Low Potential is protected with triple back to back diodes and shown. These various diode protection schemes are similar to what HP 16065A has, although they may not be required with certain LCR meters (it's always easier to remove or not install a component than add one!!).
The DC comes thru a couple resistors (10K and 100K) and a shunt external noise reduction capacitor, Diode D19 is optional to prevent reverse bias if a polarized capacitor is used (C9 and/or C11). A sense line is provided, also with an 10K and 100K resistors and shunt capacitor to allow monitoring the DUT voltage. Unipolar overvoltage monitoring is provided by Zener D17 and an LED, the Zener can be selected based upon the desired voltage range of the fixture, and if bipolar operation is required these should not be installed.
Note that the High Current series capacitor(s) are outside the High Potential Sense terminal, which also has series capacitor(s) but is a much higher impedance than the High Current terminal. The general idea is that when a "Short" calibration is performed, the calibration includes the effects of the series capacitors and reduces their measurement effects. A drawback with larger capacitors is the longer time constant to charge the DUT and the fixture capacitors, if one tends to measure larger capacitors, then reducing the 100K charging resistor might be beneficial.
One of the reasons for interest in this fixture is to help understand the subtle voltage effects on certain ceramic capacitors, and later polymer and electrolytic types. Note that the PCB is designed for all sorts of variations to allow customization as required.
Anyway, thought others might find this useful, we plan on constructing a few types with different components for various voltage ranges and capacitors types.
One word of CAUTION, this has NOT been tested on any LCR meter (yet), and could cause LCR meter damage, so please proceed with care and YMMV!!
We'll provide PCB design files and such after we've had time to evaluate the concept and verify the performance. Also likely will design a 3D printed case to hold the PCB.
Comments, suggestions, and critique are welcome.
Best,