Medical-rated toroidal isolation transformers are readily available with two 120 V primary windings and two 120 V secondary windings.
Some have 100 V taps on the primary windings.
Since these transformers are rated for 50/60 Hz, I assume that it would be safe to apply 120 V, 60 Hz to the "100 V" taps, increasing the nominal output voltage from 120 V to 144 V. 120 V at 60 Hz requires the same core flux as 100 V at 50 Hz, so the core should still be below saturation.
In this case, I also assume that the VA rating remains constant, and that therefore the secondary current ratings scale down by the same factor.
Is this correct?