Hi all,
I'm back from the holidays (actually since two weeks) and I managed to make some more tests on my Trimble GPSDO to try to tame it.
First test:
Oscillator was removed from the board, I powered it up with a clean 6V linear PSU and logged the current, just to see if the thermal regulation is good or not. The current goes up and then settles down to 450mA. It does not require much time, some minutes.
Vtune works, as oyu can see from graph. Y are values in Hz from 10MHz reference, X are Vtune values.
To my eyes, everything seems to be fine.
Second test:
I put the oscillator back in the board and powered it up, monitoring Vtune, current and frequency.
Sorry for creepy graphs, but the code was made in a hurry, during lunch
I'm still monitoring the oscillator after a couple of hours.
UCCM >SYSTem:STATus?
57964-15 serial number 86854165 firmware ver 3.0.0.11-01LINK mode
Reference Status __________________________ Reference Outputs _______________
XX Ref 8KHz 0: [LOS]
XX Ref 8KHz 1: [LOS] TFOM 7 FFOM 1
XX Ref 8KHz 2: [LOS] UCCM A Status[OCXO WARMUP]
XX Ref 8KHz 3: [LOS]
>> GPS: [phase:-1.4E-02, settling]
ACQUISITION ...................................................[GPS 1PPS Valid]
Tracking: 10 ___ Not Tracking: 1 ________ Time ____________________________
PRN El Az C/N PRN El Az GPS 13:57:56 08 Sep 2021
1 13 158 36 2 14 320
22 25 107 41 ANT DLY 0 ns
4 72 44 42 Position ________________________
3 43 95 42 MODE Hold
17 12 229 37
9 65 252 44 LAT N 43:35:11.823
19 21 251 38 LON E 13:31:1.253
26 6 64 39 HGT +225.60 m (MSL)
6 47 302 43
7 20 181 41
TFOM is still very variable, while Vtune is stuck at 1.54V, so it seems to me that no attempt is made to "tame" the oscillator.
Do you have any idea? Is the oscillatore gone? Is the board faulty? Still having a loto to test.
Thank you for your attention!