I don't really understand the two FET circuit.
I understand it's for photovoltaic mode ... but why not just capacitively couple the JFET source follower to the cathode then and avoid the ugliness trying to match current sources?
PS. I assume you are low passing before measuring the noise right?
yes, 5th order 200Hz low pass filter
See: https://www.eevblog.com/forum/projects/help-me-understand-photodiode-amplifier-jfet-bootstrap/msg5070655/#msg5070655
Thank you, good hint, I found this topic, and I modeled bootstrap in LTspice and unfortunately does not improve things as it is
.. and put your entire stuff into a metal cookie box, inclusive the batteries.. Output via a BNC connector and a shielded coax..
The noise you see may come from outer sources easily, you work with sensitive circuits..
You are right, I did exactly as you advise. I even put metal mesh in front of the photodiode, as it was picking too much 50Hz interference
A mere 1pF of parasitic capacitance in parallel with 1GΩ already adds a 1st order lowpass at 160Hz, so BW of the circuit is unlikely to be much higher than intended. But, if necessary, the above shows how BW can be readily reduced.
It's important that bootstrapping eliminates most of opamp's voltage noise only to replace it with voltage noise of the JFET source follower, which will now appear across the diode. This helps because good discretes are less noisy than chips, and you can run them at higher bias than a power-conscious IC design does. Make sure to use a low noise JFET at fairly high drain current. If you have two, maybe try connecting them in parallel instead (for lower noise) and adjust bias current (and resulting Vgs) simply by tuning the pulldown resistor.
You could replace the photodiode with an "equivalent" ordinary 1.2nF capacitor to see how much noise comes from the TIA and how much is legitimate noise from the diode.
Thank you for so much input! I measured BW and it is around 200Hz. OP amp input, photodiode and feedback resistor are soldered in air.
I replaced photodiode with capacitor and for BPW21R photodiode seems most of the noise comes from TIA, as it did not change.
Interestingly I tried same thing with TEMD5000 photodiode (same surface area, but only 70pF capacitance) and here some of the noise comes from photodiode.
I played with bootstrap circuit like on the picture below, and increasing voltage helps and eliminating one of the resistors helps as well. I am sorry, which resistor is pull down resistor here?