This is not a replacement for a "proper" noise measuring amplifier/filter as discussed in other thread(s). However I made this filter to measure and compare LF noise on HP3245A and other sources with not the lowest noise levels and it is perfectly useable for this purpose. The filter is connected to the input of a (slightly modified, hence "M") Keithley 2015M set at 1NPLC and 100mV DCV. The self-noise of Keithley with a shorted input is below 1uV p-p, with the passive filter connected it increases to about 2uV p-p - or about 0.2ppm at 10V DC input. The components are what I had at hand, a pack of 2.2uF polypropylene capacitors, a small diecast box and a 3-position switch (to defeat the filter fully or only the LPF part, the switch and the protection diodes are not shown on the circuit).
The advantages of a relatively small PP capacitor are very low leakage (picoAmps) and very fast settling time if the DC voltage is changed. The disadvantages are higher noise, first order filtering and a slight attenuation (about 2%, so can be considered benign).
Cheers
Alex