So here's a new one on me. I'm a Met/Cal procedure writer so I'm always looking for ways to remote interface with equipment to automate calibration as much as possible. Well, a Fluke 787 Processmeter crossed my desk and while I was browsing Fluke's procedure database, I came across a procedure to communicate with this unit via RS-232. So I looked the meter over and couldn't find any communication ports. I opened the battery compartment and couldn't find any pins. So both stumped and curious, I download the scripted procedure and dove into its code. In the header, it mentioned a dedicated calibration module. Because blasted Fluke changing communication interface... *@#$% fraking %## frak #$% bought enough $%@$ adapters already*
Anyway, I look up this calibration module (link below) and I discover the RS-232 cable connects to it and I figured that this must just be PC controlled switcher or something. But reading Flukes description of this module is what's really bizarre.
"Application: The calibration module provides a 9 volt dc supply to the 787. At the same time, it facilitates RS232 communication from the PC running MET/CAL to the 787 over the battery leads of the 787."
So... It's communicating over the battery leads?! How the heck does that work?? I've heard of using wall AC as a carrier wave for digital signals (poorly) but this is something completely off the maps for me. Has anyone else ever seen this? Has anyone ever successfully communicated with one of these meters before? Could I make one without giving Fluke another grand for a stupid proprietary communication adapter?
The questions aren't really all that important. I have just never seen this before and wanted to share.
Link to the calibration module:
https://www.fluke.com/en-us/product/accessories/other/fluke-787calEdit 1: Here's a link to its image as Fluke doesn't have a product picture for some reason:
https://images.app.goo.gl/KvPgjpkxxYz7AGDw7